Good day. Some feedback
What is R4 and R8 for? Seems like they will worsen DC feedback depth.
If its a 10V input, both OPA189 or OPA140 can't be powered by +12V, since the allowed common mode input range is (V+)-3.5 V (OPA140) and (V+)-2.5V (OPA189).
The bottom circuit seems to be a composite amplifier where main purpose of U2 is to make it so U1 doesn't dissipate power.
It should be possible to make the DC servo stable and without peaking in frequency response by playing with opamp speed and output RC if that is what you mean by stability and flatness.
The chopper "scary thing" are current pulses at the inputs, not on the output. Check the ADA4523 datasheet current noise paragraph
Also check stability with load capacitance, since for a DC reference there is likely to be capacitance somewhere. Its not a bad idea to put some capacitance on the output too
What are your requirements for this circuit? I'd work from those and try the simplest circuit that would meet them
Thanks for the replay.
Maybe i was not clear from previous post this is only simulation not real design. Whit this in mind here is my answers.
- R4 and R8 are used to sasisfy the simulator - to find initial dc point. 100M shoud not influence simulation. in real design they are not present.
- +-12V normaly i use 7V but yes in real design I use +-15V for other reason.
For clarity, I will show the actual design used on my current Vref board.
Short explanation: The outputs of U8A and U8B are used as feedback/monitor to assess the confidence interval of the voltage transfer. For example, the charge-injection errors of the Keithley DMM6500 and the two HP 3456A units are close to the noise level. The DMM6500 performs slightly better. The HP 34401A is not as good, but its performance is still acceptable. The Pico M2000 performs worse. Some flukes are bad too while Brymen 869s is beter than any fluke
Lessons learned:
I previously used Vrefs with a simple, classical buffer consisting of an op-amp and an NPN transistor. At that time, I did not know what I was actually measuring, so I was 100% confident.
Attaching handheld multimeter At U8B output and measuring AC and DC part give reasonable good result, but attaching of active high speed probe and 12-bit scope is more fun.
The most important lesson is:
The less you know, the fewer problems you see. The more you know, the less confident you become.
My questions
- which circuit you will choose and why?
- for which application you will choose the dual loop buffer and when servo DC?
I dont need to tune existing circuit. I need good fundamental block E.g. Jacque give me very valuable answer.