Further to posts #79 and #83 by eppley, I contacted the museum and Barry M responded with some history, advice and photos of the relevant Eppley factory records for a 'MIN 1' model standard cell that is in an RFL model 829G AC/DC Calibration Standard I am part way through restoring.
The Eppley records confirm the calibrator was manufactured circa late 1974 to early 1975, which is also confirmed by datecodes on some internal parts. One could suggest from those records that the batch of 25 cells ordered by RFL were for a nominal month of calibrator production, and further suggest that about 500 calibrators per annum were being produced, as the cell was an optional extra. That sort of aligns with serial number information in the 829G manual, so a customer like RFL could well account for circa 1500 cells for a product line over 5 years.
The museum's donation of Eppley records involves about 20 shelves of three-ring binders and covers about a million cells, so that makes RFL just a minor user, given that Barry indicated they made circa 20-30 cells per day (eg. 6,500 per annum ballpark).
The cell in the calibrator is still functional at about 1.01811V, indicating a nominal 24uV loss per year, which is consistent with the 20-40uV/yr advice in NBL Monograph 84 from 1965, and is not behaving erratically so far. The cell can stay in the calibrator, which luckily includes terminals for an external reference, and enjoy its retirement now that it has a copy of its birth certificate.