Author Topic: VB .NET 2019 and NI-VISA Example Needed  (Read 12042 times)
Offline NY2KWTopic starter
« on: April 10, 2020, 03:12:21 am »
I am teaching myself VS 2019 Community with VB .NET using NI-VISA to communicate with a Siglent device.  All of the examples I can find are in C# or old VB6.  I am having difficultty with the NI-VISA ResourceManager class.  I have found older examples that use
Public Class Form1
    Dim result As String
    Dim dut_tcpip As TcpipSession
    Dim visa_string As String
    Private Sub Button1_Click(sender As Object, e As EventArgs) Handles Button1.Click
        visa_string = TextBox1.Text
        dut_tcpip = ResourceManager.GetLocalManager.Open(visa_string)
        dut_write("*IDN?")
        dut_read()
        MsgBox(result)
    End Sub
The problem is the latest version of NI-VISA ResourceManager does not list GetLocalManager.  I have searched through current and previous NI-VISA documents and cannot find how this changed or to what.  Can someone help me get past this first door?
Thank you in advance and stay safe and healthy
Offline BroMarduk
Re: VB .NET 2019 and NI-VISA Example Needed
« Reply #1 on: April 10, 2020, 03:27:15 am »
Just a quick check...do you have:
Imports NationalInstruments.VisaNS
at the top of the file?
Then try
Public Class Form1
    Dim mbSession As MessageBasedSession
        mbSession = CType(ResourceManager.GetLocalManager().Open(visa_string ), MessageBasedSession)
        Dim responseString As String = mbSession.Query("*IDN?")
        MsgBox(responseString)
        mbSession.Dispose()
    End Sub
From the documentation here:  https://www.ni.com/en-us/support/documentation/supplemental/17/archived--c--and-visual-basic--net-instrument-control.html
You can then use the mbSession object to transfer data and get the response.
** NOTE  I realized this is from the old library....Let me see if I can find the new one **
« Last Edit: April 10, 2020, 03:40:17 am by BroMarduk »
« Reply #2 on: April 10, 2020, 03:37:50 am »
Thanks for the reply.  VisaNS is no longer in the current NI-VISA download.  I believe the VisaNS version had  'GetLocalManager' under ResourceManager.   I have added the new NI version under References in Visual Studio and in the code I have:
Imports Ivi.Visa
I cannot find any examples in VB .NET with the new version of NI-VISA Resource Manager... I have Googled literally for a few hours looking
Offline BroMarduk
« Reply #3 on: April 10, 2020, 03:40:46 am »
LOL... I caught that a bit late.
Let me see...
Offline BroMarduk
« Reply #4 on: April 10, 2020, 03:41:24 am »
Do you have a C# example I can translate?
« Reply #5 on: April 10, 2020, 03:50:34 am »
First, Many many thanks for your kind help.  Yes, I think I can find a C example from NI-VISA.  Gi8ve me a few minutes. In the meantime I did your code:
Imports Ivi.Visa
Public Class Form1
        mbSession = CType(ResourceManager.GetLocalManager().Open(visa_string), MessageBasedSession)
        mbSession.Dispose()
    End Sub
End Class
I get errors that GetLocalManager is not a member of ResourceManager and that Query is not a member of mbSession.
Let me get that C example
« Reply #6 on: April 10, 2020, 03:58:23 am »
Here is a read/write example from NI-VISA with C NET.  Its probably a lot more than you need
using System;
using System.Windows.Forms;
using NationalInstruments.Visa;
namespace NationalInstruments.Examples.SimpleReadWrite
    /// <summary>
    /// Summary description for Form1.
    /// </summary>
    public class MainForm : System.Windows.Forms.Form
        private MessageBasedSession mbSession;
        private string lastResourceString = null;
        private System.Windows.Forms.TextBox writeTextBox;
        private System.Windows.Forms.TextBox readTextBox;
        private System.Windows.Forms.Button queryButton;
        private System.Windows.Forms.Button writeButton;
        private System.Windows.Forms.Button readButton;
        private System.Windows.Forms.Button openSessionButton;
        private System.Windows.Forms.Button clearButton;
        private System.Windows.Forms.Button closeSessionButton;
        private System.Windows.Forms.Label stringToWriteLabel;
        private System.Windows.Forms.Label stringToReadLabel;
        /// <summary>
        /// Required designer variable.
        /// </summary>
        private System.ComponentModel.Container components = null;
        public MainForm()
            // Required for Windows Form Designer support
            InitializeComponent();
            SetupControlState(false);
        /// <summary>
        /// Clean up any resources being used.
        /// </summary>
        protected override void Dispose( bool disposing )
            if( disposing )
                if(mbSession != null)
                    mbSession.Dispose();
                if (components != null)
                    components.Dispose();
            base.Dispose( disposing );
        #region Windows Form Designer generated code
        /// <summary>
        /// Required method for Designer support - do not modify
        /// the contents of this method with the code editor.
        /// </summary>
        private void InitializeComponent()
            System.ComponentModel.ComponentResourceManager resources = new System.ComponentModel.ComponentResourceManager(typeof(MainForm));
            this.queryButton = new System.Windows.Forms.Button();
            this.writeButton = new System.Windows.Forms.Button();
            this.readButton = new System.Windows.Forms.Button();
            this.openSessionButton = new System.Windows.Forms.Button();
            this.writeTextBox = new System.Windows.Forms.TextBox();
            this.readTextBox = new System.Windows.Forms.TextBox();
            this.clearButton = new System.Windows.Forms.Button();
            this.closeSessionButton = new System.Windows.Forms.Button();
            this.stringToWriteLabel = new System.Windows.Forms.Label();
            this.stringToReadLabel = new System.Windows.Forms.Label();
            this.SuspendLayout();
            // queryButton
            this.queryButton.Location = new System.Drawing.Point(5, 83);
            this.queryButton.Name = "queryButton";
            this.queryButton.Size = new System.Drawing.Size(74, 23);
            this.queryButton.TabIndex = 3;
            this.queryButton.Text = "Query";
            this.queryButton.Click += new System.EventHandler(this.query_Click);
            // writeButton
            this.writeButton.Location = new System.Drawing.Point(79, 83);
            this.writeButton.Name = "writeButton";
            this.writeButton.Size = new System.Drawing.Size(74, 23);
            this.writeButton.TabIndex = 4;
            this.writeButton.Text = "Write";
            this.writeButton.Click += new System.EventHandler(this.write_Click);
            // readButton
            this.readButton.Location = new System.Drawing.Point(153, 83);
            this.readButton.Name = "readButton";
            this.readButton.Size = new System.Drawing.Size(74, 23);
            this.readButton.TabIndex = 5;
            this.readButton.Text = "Read";
            this.readButton.Click += new System.EventHandler(this.read_Click);
            // openSessionButton
            this.openSessionButton.Location = new System.Drawing.Point(5, 5);
            this.openSessionButton.Name = "openSessionButton";
            this.openSessionButton.Size = new System.Drawing.Size(92, 22);
            this.openSessionButton.TabIndex = 0;
            this.openSessionButton.Text = "Open Session";
            this.openSessionButton.Click += new System.EventHandler(this.openSession_Click);
            // writeTextBox
            this.writeTextBox.Anchor = ((System.Windows.Forms.AnchorStyles)(((System.Windows.Forms.AnchorStyles.Top | System.Windows.Forms.AnchorStyles.Left)
            | System.Windows.Forms.AnchorStyles.Right)));
            this.writeTextBox.Location = new System.Drawing.Point(5, 54);
            this.writeTextBox.Name = "writeTextBox";
            this.writeTextBox.Size = new System.Drawing.Size(275, 20);
            this.writeTextBox.TabIndex = 2;
            this.writeTextBox.Text = "*IDN?\\n";
            // readTextBox
            this.readTextBox.Anchor = ((System.Windows.Forms.AnchorStyles)((((System.Windows.Forms.AnchorStyles.Top | System.Windows.Forms.AnchorStyles.Bottom)
            this.readTextBox.Location = new System.Drawing.Point(5, 136);
            this.readTextBox.Multiline = true;
            this.readTextBox.Name = "readTextBox";
            this.readTextBox.ReadOnly = true;
            this.readTextBox.ScrollBars = System.Windows.Forms.ScrollBars.Vertical;
            this.readTextBox.Size = new System.Drawing.Size(275, 119);
            this.readTextBox.TabIndex = 6;
            this.readTextBox.TabStop = false;
            // clearButton
            this.clearButton.Anchor = ((System.Windows.Forms.AnchorStyles)(((System.Windows.Forms.AnchorStyles.Bottom | System.Windows.Forms.AnchorStyles.Left)
            this.clearButton.Location = new System.Drawing.Point(6, 257);
            this.clearButton.Name = "clearButton";
            this.clearButton.Size = new System.Drawing.Size(275, 24);
            this.clearButton.TabIndex = 7;
            this.clearButton.Text = "Clear";
            this.clearButton.Click += new System.EventHandler(this.clear_Click);
            // closeSessionButton
            this.closeSessionButton.Location = new System.Drawing.Point(97, 5);
            this.closeSessionButton.Name = "closeSessionButton";
            this.closeSessionButton.Size = new System.Drawing.Size(92, 22);
            this.closeSessionButton.TabIndex = 1;
            this.closeSessionButton.Text = "Close Session";
            this.closeSessionButton.Click += new System.EventHandler(this.closeSession_Click);
            // stringToWriteLabel
            this.stringToWriteLabel.Location = new System.Drawing.Point(5, 34);
            this.stringToWriteLabel.Name = "stringToWriteLabel";
            this.stringToWriteLabel.Size = new System.Drawing.Size(91, 14);
            this.stringToWriteLabel.TabIndex = 8;
            this.stringToWriteLabel.Text = "String to Write:";
            // stringToReadLabel
            this.stringToReadLabel.Location = new System.Drawing.Point(5, 117);
            this.stringToReadLabel.Name = "stringToReadLabel";
            this.stringToReadLabel.Size = new System.Drawing.Size(101, 15);
            this.stringToReadLabel.TabIndex = 9;
            this.stringToReadLabel.Text = "String Read:";
            // MainForm
            this.AutoScaleBaseSize = new System.Drawing.Size(5, 13);
            this.ClientSize = new System.Drawing.Size(287, 289);
            this.Controls.Add(this.stringToReadLabel);
            this.Controls.Add(this.stringToWriteLabel);
            this.Controls.Add(this.closeSessionButton);
            this.Controls.Add(this.clearButton);
            this.Controls.Add(this.readTextBox);
            this.Controls.Add(this.writeTextBox);
            this.Controls.Add(this.openSessionButton);
            this.Controls.Add(this.readButton);
            this.Controls.Add(this.writeButton);
            this.Controls.Add(this.queryButton);
            this.Icon = ((System.Drawing.Icon)(resources.GetObject("$this.Icon")));
            this.MaximizeBox = false;
            this.MinimumSize = new System.Drawing.Size(295, 316);
            this.Name = "MainForm";
            this.StartPosition = System.Windows.Forms.FormStartPosition.CenterScreen;
            this.Text = "Simple Read/Write";
            this.ResumeLayout(false);
            this.PerformLayout();
        #endregion
        /// <summary>
        /// The main entry point for the application.
        /// </summary>
        [STAThread]
        static void Main()
            Application.Run(new MainForm());
        private void openSession_Click(object sender, System.EventArgs e)
            using (SelectResource sr = new SelectResource())
                if(lastResourceString != null)
                    sr.ResourceName = lastResourceString;
                DialogResult result = sr.ShowDialog(this);
                if(result == DialogResult.OK)
                    lastResourceString = sr.ResourceName;
                    Cursor.Current = Cursors.WaitCursor;
                    using (var rmSession = new ResourceManager())
                        try
                            mbSession = (MessageBasedSession)rmSession.Open(sr.ResourceName);
                            SetupControlState(true);
                        catch (InvalidCastException)
                            MessageBox.Show("Resource selected must be a message-based session");
                        catch (Exception exp)
                            MessageBox.Show(exp.Message);
                        finally
                            Cursor.Current = Cursors.Default;
        private void closeSession_Click(object sender, System.EventArgs e)
        private void query_Click(object sender, System.EventArgs e)
            readTextBox.Text = String.Empty;
            try
                string textToWrite = ReplaceCommonEscapeSequences(writeTextBox.Text);
                mbSession.RawIO.Write(textToWrite);
                readTextBox.Text = InsertCommonEscapeSequences(mbSession.RawIO.ReadString());
            catch(Exception exp)
            finally
        private void write_Click(object sender, System.EventArgs e)
            try
        private void read_Click(object sender, System.EventArgs e)
            try
            finally
        private void clear_Click(object sender, System.EventArgs e)
        private void SetupControlState(bool isSessionOpen)
            openSessionButton.Enabled = !isSessionOpen;
            closeSessionButton.Enabled = isSessionOpen;
            queryButton.Enabled = isSessionOpen;
            writeButton.Enabled = isSessionOpen;
            readButton.Enabled = isSessionOpen;
            writeTextBox.Enabled = isSessionOpen;
            clearButton.Enabled = isSessionOpen;
            if(isSessionOpen)
                writeTextBox.Focus();
        private string ReplaceCommonEscapeSequences(string s)
            return s.Replace("\\n", "\n").Replace("\\r", "\r");
        private string InsertCommonEscapeSequences(string s)
            return s.Replace("\n", "\\n").Replace("\r", "\\r");
Offline BroMarduk
« Reply #7 on: April 10, 2020, 04:05:08 am »
Try it without the GetLocalManager.:
Imports Ivi.Visa
Public Class Form1
        dut_tcpip = ResourceManager.Open(visa_string)
        dut_tcpip.RawIO.Write("*IDN?" + vbLf)
        MsgBox(dut_tcpip.RawIO.ReadString())
    End Sub
« Last Edit: April 10, 2020, 04:15:25 am by BroMarduk »
Offline BroMarduk
« Reply #8 on: April 10, 2020, 04:08:31 am »
The new line character may be wrong...You can try it without the + VbLf as well.
« Reply #9 on: April 10, 2020, 04:21:21 am »
I tried it and getting an error for line:
Error says "Reference to a non-shared member requires an object reference"
Offline BroMarduk
« Reply #10 on: April 10, 2020, 05:07:09 am »
Looks like you have to create an instance of ResourceManager class:
Imports Ivi.Visa
Public Class Form1
        Using rm As New ResourceManager
                dut_tcpip = rm.Open(visa_string)
        End Using
    End Sub
« Reply #11 on: April 10, 2020, 01:02:16 pm »
No errors building but getting exception on the Open line.  Reading NI-VISA changes from the now legacy VisaNS:
However, because the new API implements interfaces specified by the IVI specification and uses several types defined in that specification, the actual members of these Session-derived classes are different from those in VisaNS and therefore will require manual changes to your applications to migrate from VisaNS to NI-VISA .NET. In some cases, such as the formatted I/O functionality, the API is significantly different from VisaNS and may require larger changes for migration.
Some features supported by the VisaNS API are not supported or are supported differently in the NI-VISA .NET API. Some of the major differences are described below.
Message-based I/O operations are provided in a different way
Simple message-based operations, such as ReadString or Write, that are implemented by some Session-derived classes in VisaNS are now provided by the RawIO property of the Ivi.Visa. IMessageBasedSession interface that some Session-derived classes implement. The RawIO property is of a type that implements the Ivi.Visa.IMessageBasedRawIO interface.
Similarly, the formatted message-based operations that are implemented by MessageBasedSessionReader and MessageBasedSessionWriter in VisaNS are now provided by the FormattedIO property of the Ivi.Visa.IMessageBasedSession interface that some Session-derived classes implement. The FormattedIO property is of a type that implements the Ivi.Visa.IMessageBasedFormattedIO interface.
Format specifiers used for formatted I/O are different
The formatted I/O methods in VisaNS support format strings similar to those used by the String.Format method. NI-VISA .NET, however, supports format strings with format specifiers similar to those used by VISA viPrintf and viScanf functions. For more details about the FormattedIO methods and supported format specifiers, see the description of Ivi.Visa.IMessageBasedFormattedIO interface in the VPP-4.3.6 specification.
ResourceManager class is disposable
The ResourceManager class is used differently. To get an instance of a ResourceManager object, construct a new instance using the public constructor ResourceManager(). The ResourceManager object is disposable and should be disposed once no longer needed. Note that any Session-derived object obtained by using the Open method of a ResourceManager remains valid after the ResourceManager object is disposed.
You would think that NI would provide at least some examples of code with these changes.
« Reply #12 on: April 10, 2020, 02:56:08 pm »
Thank you again for all your help.  I found an obscure thread in an NI Forum on Instrument control discussing the changes to ResourceManager from legacy VisaNS to NI-VISA.   You were 100% in your suggestions.  In console style program, the following returns the Instrument type, name and serial number in a Command window.  Now moving to Windows Form
Imports System
Imports System.Collections.Generic
Imports Ivi.Visa
Module Module1
   Dim visa_string = "USB?*"
   Sub Main()
        Using rm = New ResourceManager()
            Try
                Dim resources = rm.Find(visa_string)
                'Dim resources As IEnumerable(Of String) = rm.Find(visa_string)
                For Each s As String In resources
                    Dim parseResult As ParseResult = rm.Parse(s)
                    System.Console.WriteLine(s, parseResult.InterfaceType)
                Next
            Catch ex As Exception
                System.Console.WriteLine("Exception" + ex.Message)
            End Try
        End Using
        System.Console.ReadLine()
    End Sub
I didn't understand the commented line why IEnumerable was used so I simplified it and still worked fine.   Now onto using VS to put this and more into a Windows Form app.
Will keep you posted.
Offline BroMarduk
« Reply #13 on: April 10, 2020, 09:00:24 pm »
Glad you found the solution.  It's hard debugging an API you don't know in your head (but the basics still applied).