Hi,
Did someone say more pictures?
Let me see if I can take care of that with some meaningful information in it.
Small corection schematic calibratorFirst, a slight modification of the calibration schematic.
I didn't think it was really neat of me that I hadn't put a little more effort into bringing the duty cycle more to 50% without increasing the number of components.
I used a 555 calcualtor that was available online, to maintain reasonable resistance values and bring the duty clycle as close to 50 percent as possible.
Also, the I wanted to use the component values from the E12 series, so no extraneous values.
I have, using the "on line" calculator
was able to meet the conditions of standard E12 values and finally achieve a duty cycle of 50.25%.
What I pointed out earlier, the 2nd, 4th, 6th, etc harmonics disappear from the FFT at exactly 50% duty cycle.
The reason I finally made it a little better is because I could see on the scope that it was not a nice sine wave.
And now I can no longer see that, the distortion is now just over 2% and was about 3.3%.
But for level reading this adjustment was/is not necessary.
Disadvantages of this modification, the only thing is that due to the lower value R3 is now 1K2 and therefore the current consumption is just over 3mA, for me this is not a problem.
This is because this circuit is only on for a short time anyway.
The modified calibrator schematic for lower THD.
In this FFT image it can be seen that the even harmonics are now much lower in strength this compared to the previous FFT image of the calbrator.
First set of scope pictures when the 100x amplifier is used.I use the Hameg and one of my Siglent scopes for the following measurements.
The Hameg scope is used to clarify something faster, this is because this scope has e.g. a "Quick View" button and sometimes filtering can be set faster.
These kinds of measurements I show here take a lot of time to make,
if you want to explain something you have to make sure you do it carefully and measurement errors are easily made.
The scoops used are always set to 20MHZ bandwidth.
The input impedance always set to 1Meg, this is because the measuring amplifier is designed for that.
If filtering or averaging has been applied, this is indicated.
This measurement was done with an open input of the 100x amplifier.No filtering or averaging was applied and the scope was set to "Peak & Hold"
Now the input of the amplifier is terminated with a 50 Ohm termination resistor.Note the vertical scale, it is now 10uV/Div and in the previous picture the sensitivity was 50uV/Div.
This is a nice check if you have an amplifier that has low noise., input is shorted.The noise in this measurement is slightly lower than the measurement with 50 Ohm termination resistor.
Scope noisefloorTo show that the noise of the scope is low enough I left everything the same, just removed the 100x amplifier and terminated the scope input with 50 Ohms.
The noise level of the Siglent scoops I also used are lower than the Hameg scope.
Measurement amplifier performance with different signals.Here it is clearly visible that the measuring amplifier has a large bandwidth, it is not visible here at a 1KHz square wave that the measuring amplifier is between the Generator and the scope.
Only at the rise time is visible that this does not quite match what the Generator provides in terms of rise time.
Also remember that the scope is at 20MHZ bandwidth.
1mV input wide band, so no filters1mV input 100KHz Low passThis is with the same input signal only a 100KHz filter at the output.
1mV input 20KHz Low passAnd now with the 20KHz filter on the output of the measuring amplifier.
The signal got a little smaller, which is partly due to the lower bandwidth, but also partly due to the filter used.
The component values of the filter used are slightly different from those shown in the amplifier schematic, but the filter has the same bandwidth.
So only the impedances are a little different, which is otherwise not very important here.
100uV input, No filtersHere one picture at 0.1mV input signal and then without any filters except the scope 20MHz filter.
The broadband noise is now well visible at this low signal level.
But keep in mind, that the measurement amplifier has a -3dB bandwidth of 1MHz.
10uV input, No filtersNow we go one step lower with the input signal, and this picture was taken with 10uV top top and across the full bandwidth of the measuring amplifier.
There is very much noise visible now, and at this bandwidth, at this signal level, little sense will be made of what appears on the scope image.
10uV input, 100KHz Low PassThe signal can be recognized quite well when the bandwidth is reduced to 100KHz.
10uV input, 100KHz Low Pass + 1024x Avarage1024 times means is naturally exaggerated and it also takes a lot of time to build this scope picture.
But it shows well what is possible with a modern scope.
10uV input, 20KHz Low PassBy using the 20KHz filter, it is also quite possible to get a good picture of this small signal.
The first measurements on a LAB power supply!But doing measurements involves something beforehand, and that is the proper way to connect the measurement cable to the Power Supply.
But first a few pictures of the cable I use for this.
That is a coax cable from Belden RG58 with the type number: 9223
And in this way I prepared the cable for measurements on lab power supplies.
I added a little solder to both ends of the cable.
And I pinched the shield a little flat with the pliers visible in the foreground, this to secure the cable under the terminal blocks/connectors (if any)
It is of course also possible to use small connectors that you can clip under the banana connectors.
Keep in mind to keep the measuring cable a little longer, but not too long, because of EMC signals.
It is best to take the length into account because this cable is not very sturdy for the way I connect it.
This is the right way with these banana connectors that most power supplies are still equipped with.This picture shows what is almost the best way to connect the measurement cable.
At a 90 degree angle with the twisted banana wires running to the load.
And so connect the measurement cable in the way shown in the photo!
I have seen many videos on "junk tube" of people testing power supplies and dynamic loads not knowing how to do it properly.
So a few more pictures on how NOT to connect the test lead. :-)
Seems convenient, but it is not the right way, think about connector resistance and inductance of the connections, especially if you're going to do dynamic testing.
Haha, You are a comedian by profession? And now some real measurements on an HP 6237A power supply.And first, the +output which can go from 0 to +20V and can be loaded with a maximum of 500mA.
The output is loaded with about 300mA in this measurement.
I show measurements without filtering, at 100KHz bandwidth and 20KHz bandwidth.
This is a measurement with 100KHz bandwidth.This is a measurement with 20KHz bandwidth.I draw the conclusion that this channel is nice and clean.
Only a very small portion of the noise at higher bandwidth comes from the power supply itself.
I have done tests with and without load on the power supply and the differences in noise are very low, otherwise not worth pictures.
But not everything is so nice about this power supply, I now let you see a measurement of the 18V channel.
18V @ 1-Ampere 100KHz bandwidth.At the 18V output, the mains frequency is doninand.
But with 0.8mV top top, this power supply still meets modern requirements.
The other measured channel with 100mV top top is really very good.
More More MoreBy the way, there are many more measurements to do and a few I did, but what you see here today in this topic, is 8 hours of work. :-)
I want to tell more about wiring, so how to connect things and what to look out for when doing P.A.R.D measurements.
Just connect a second measurement cable, a coaxial cable to trigger something and you suddenly find yourself with a completely different signal on your scope, why?
You think you are wise and you connect the ground connection of your power supply to ground of your measurement system and you have a horror signal on the scoop, why?
cell phones of your own and/or colleagues, transmitters in the neighborhood, etc, etc.
So many things to tell, to make as few measurement errors as possible.
And of course, what is of importance of the PARD signal you see on the scope.
The noise up to 100KHz is probably local filtered by decoupling in the D.U. T. you are testing.
Connecting cables that are too long and not twisted are wonderful antennas for both your power supply, but also for your D.U.T.
And I now only showd measurements of linear power supplies.
The fun starts when you start measuring on anything related to SMPS power supplies.
I'll will also show some picture's of an SMPS power supply in a future post and I'll call this power supply an EMC Lighthouse.
Time to make dinner for my Girlfriend!
Shoot!
Bram