Even if the 7107 is sensitive to heat, its to 100uV tops, so unless your range resolves 0.1mV it may not be a consideration.
If you don't used matched resistors for tempco and other features, then even if they are precision they will drift at different rates.
If designs internal temperature varies at maximum warmup, it will still drift the 7106, so what matters is to carefully consider operating temperature against the components thermal stability and drift. As the number of digits in your panel meter increase, the more all these small factors will matter; if you are just doing 1% for a 3.5 digit DVM, most resistors and precision in the 1% or less will do, including non-matched pairs, and they use of either 7107 or 7106 is more the visual appeal of LED vs LCD than a question of accuracy.