Attached is an example of the phase delta I see. I've been playing around with this for years, finally thought I would figure it out. I picked up a 5370B recently and I am using that as the TIC with an offset of 1hz in this run. These aren't phase wraps, I'm missing a very basic point I think. I'm thinking of swapping in TUF3 mixers for the SYPD-1 phase detectors. I thought phase detectors and mixers behaved the same when close in frequency but have never researched it.
Thanks.
you can also see the subharmonics in the trace. I thought it might be a reflection, so I tried resistive as well as reactive power splitters. I also tried just a 'T' without luck. I also put directional couplers on the inputs at one point, 10Mhz filters, tried a TICC instead of the 5370, used a DDS and a 3336 and 3325 synthesized generators for the offset, changed the power and am now using lt3045s and lt3094s to feed the DMTD board. I tried QBit low noise amps ahead of the splitter (splits the reference to reference and DUT for testing) and then used connectorized attenuators to balance the signals. I brought the phase into alignment several ways but since the phase varies, that didnt help, all it did was start the curve in a different quadrant. I thought I had it when I found an opamp with excessive floating input voltage. I also thought I had it when a coupling cap was in backwards. Nope and nope.
This board used transformers to couple in the signals. I had to make some changes to ensure the load was the same on all inputs (after I found the problem).
I put probes on the 5370B and tested it at different points and the offset delta swing starts post phase detector. I also used an HP 4815 vector impedance meter to probe the signal path of the reference vs the DUT and they were near perfect. By the way, that is my favorite piece of test equipment, the 4815. Mine is perfect in every way. I have a parts unit as well. Bought both for $80 bucks...
Thanks again.