A 5nH measurement isn't really believable with such an instrument, at any test frequency.
And what's the minimum inductance that LCR tweezers can or should measure?
As we've tried to help you over and over again, as many others have, low Z and high Z components are difficult with tweezers. The tweezer data sheet should show a error graph like Grandchuck has shown, or at minimum should state the lower and upper ranges of measurement.
5nH inductive reactance at 100kHz is 0.003Ω, at 10kHz it's 0.0003Ω, at 1kHz it's 0.000003Ω

Would think a reasonable measurement Z range for quality tweezers and measurements would be 0.1Ω and above with careful attention to detail, this equates to 160nH @ 100kHz, 1.6uH @ 10kHz and 16uH @ 1kHz.
You always bring a

when you show the 5nH Inductance Tweezer measurement with 100 femto Henry resolution (yes that's femto Henries!!), see post #18. Here's the image

Please spend some
serious time studying the physics of components, the methods of measurement, the instruments and the range of such. Then and apparently only then will you begin to understand the ridiculous tweezer measurements you keep claiming. Listen to what folks here are trying to say and help you with, instead of ignoring such!!
Best