Crosspost from a TE thread from earlier...
I Finally got around to building my new uDR1399 reference boards. They use bootstrapping and an LT5400 to set the amplifier gain. Both of those are things I would have wanted to see in the already available standards like the TEN-399 board and VREF10. I may be incorrect on what the VREF10 uses, so feel free to correct me. They have only burned in for about 72hrs so far, but seem to be humming along stabily. The results are from unit 1 of 3 that I have built.
I will edit and reiterate what is mentionned on the GitHub docs in spite of issues. As of current, do NOT connect a battery or other live supply to the unit while making connections. I tend to use a linear power supply and use the ON / OFF control of the channel to turn the uDR1399 unit on after connecting. Please add a 47uF electrolytic across the input posts if you wish to do so. There is a known issue where transients resulting from the hotplugging of the power cables can cause transients exceeding the LT3045's maximum inputs.Unit 2 is reserved for Josh, but unit 3 can be spoken for if someone wishes to test a reference like this.
I measured using my DMM6500 on 10NPLC, 10 sample averaging DCV filter, AZ on, and 10GOhm InZ. Sensing is done through Banana-BNC adapters, and the internal wiring in my refs are twisted pair to the Pomona 3770 plugs. The STD-Dev running is using 1799 samples (or what 10NPLC for 30s would work out to, not including averaging filter) as suggested by iMO. The noise is high compared to the 34465A, but I have gotten ~4.15uV STD-Dev when disabling AZ, which is contrary to results we've seen so far in this thread.
More info can be found on my GitHub where I documented everything about the project:
https://github.com/T-622/uDR1399-10V-Reference-Standard/tree/mainStability test results can be found here:
https://www.eevblog.com/forum/testgear/bench-dmm-stability-comparisons-with-testcontroller/Josh has one of the reference boards, and is testing it with a Keithley 2010, 34465A, and DAQ6510. He has reported numerous sets of findings, highlighting the temperature stability of the boards, and the overall noise within 30min windows on a Keithley 2010 to be <1uV std-dev. I am also attempting to build a thermal chamber to test the temperature coeffecient of the voltage output. This is proving quite difficult.
Tests Conducted:
1.) Josh's Initial Tests:
https://www.eevblog.com/forum/testgear/bench-dmm-stability-comparisons-with-testcontroller/msg6332874/#msg6332874Results Using 34465A For Best Stability:
Moving Filter:
2min - 10 Sample Moving Filter: 582nV std-dev
50min - 10 Sample Moving Filter: 727nV std-dev
75min - 10 Sample Moving Filter: 753nV std-dev
100min - 10 Sample Moving Filter: 751nV std-dev
No Moving Filter?:
018min - 2.92445uV std-dev
030min - 3.02086uV std-dev
040min - 2.90277uV std-dev
075min - 2.63610uV std-dev
100min - 2.46730uV std-dev
2.) Josh's Continued Test of 10 Sample Moving Filter:
https://www.eevblog.com/forum/testgear/bench-dmm-stability-comparisons-with-testcontroller/msg6333148/#msg6333148Using 34465A:
* Note: These sets of tests featured a lot of thermal variance in the uDR1399's environment temperature, and the thermal influence on the reference was relatively minimal.
075min - 10 Sample Moving Filter: 785nV std-dev
128min - 10 Sample Moving Filter: 817nV std-dev
267min - 10 Sample Moving Filter: 1.61831uV std-dev
360min - 10 Sample Moving Filter: 2.04978uV std-dev
500min - 10 Sample Moving Filter: 2.92553uV std-dev
3.) Battery Bank + Keithley 2750:
https://www.eevblog.com/forum/testgear/bench-dmm-stability-comparisons-with-testcontroller/msg6333938/#msg6333938No Filter:
40min warmup: 2.56582uV std-dev. I wouldn't count this one, as the reference was still warming up. However, it appeared to reach relative stability within a 16.67min window.
Test 4: 2min Filter ON: 811uV std-dev
Test 5: 2min Filter ON: 866nV std-dev
Test 6: Filter is 10 sample moving average (?) 1799 sample std-dev moving used in TestController.
2min - Filter ON: 347nV std-dev
10min - Filter ON: 677nV std-dev
30min - Filter ON: 848nV std-dev
100min - Filter ON: 668nV std-dev
4.) Tyler - DMM6500. 10 Sample moving avg filter, AZ ON, 10V range, 10PLC:
https://www.eevblog.com/forum/testgear/bench-dmm-stability-comparisons-with-testcontroller/msg6335186/#msg6335186* Note: My reference featured here has been burning in for 432hrs at this point of measurement.
140min - Filter On: 2.906uV std-dev. Very little temperature variation
5.) Josh - Keithley 2010 Test:
https://www.eevblog.com/forum/testgear/bench-dmm-stability-comparisons-with-testcontroller/msg6335736/#msg6335736Test 1: 10PLC, 10 sample filter enabled, 1799 sample std-dev-running enabled.
2min - 612nV std-dev
5min - 654nV std-dev
10min - 677nV std-dev
638min - 875nV std-dev. 2010 Std-dev as measured by TC is 3.1577428uV.
Test 2: 10PLC, 10 sample filter enabled, 1799 sample std-dev-running enabled.
5min - 1.2195uV std-dev
Test 3:
5min - 803nV std-dev
10min - 671nV std-dev
15min - 711nV std-dev
Tyler