Test Equipment / Re: New 2ch pocket DSO+SG - Zeeweii DSO2512G
« Last post by Sham73 on Today at 09:33:38 pm »**Open user calibration function.What sources do you use for AC and DC calibration?
How to perform AC Gain calibration:
1. The oscilloscope measures the 2.5V 1Khz sin waveform. Observe the "PKPK" measurement item.
2. Open Menu-> Meas, then press the CH1CH2 button, and then long press the [mV] key to enter the AC Gain calibration mode.
3. The UP DOWN button is used to increase/decrease the CH1 Gain until the "PKPK" voltage is equal to or close to the voltage of the signal generator.
4. The LEFT RIGHT button is used to increase/decrease the CH2 Gain until the "PKPK" voltage is equal to or close to the voltage of the signal generator.
5. Press the OK button to exit and complete the calibration.
How to perform DC offset calibration:
1. The oscilloscope measures the DC power supply. Observe the "Mean" measurement item.
2. Open Menu-> Meas, then press the CH1CH2 button, and then long press the [V] key to enter the DC offset calibration mode.
3. The UP / DOWN button is used to increase/decrease the CH1 dc offset until the "Mean" voltage is equal to or close to the voltage of the multimeter.
4. The LEFT / RIGHT button is used to increase/decrease the CH2 dc offset until the "Mean" voltage is equal to or close to the voltage of the multimeter.
5. Press the OK button to exit and complete the calibration.
Different vertical sensitivities use different calibration parameters, so they need to be calibrated separately.
**add [ZOOM]
Long press the "s" key to turn on/off Zoom mode.
(Zoom is not supported when roll mode or cursor is on\
It is recommended to exit Zoom mode before shutting down to avoid measure item are cleared)