Author Topic: Data Precision 8200 - 100V range out of pot range during adjustment  (Read 948 times)
Online RaxTopic starter
Hi all,
I am adjusting a DP8200 and everything is going alright, except the 100V range is not adjusting enough to get into calibration. I am barely at 69.9038V and I've spun the pot to much I am certain I'll run out of range before I get it within 100uV of 70V as it's supposed to be.
I enclose the schematic of the corresponding circuitry.
Any input and thoughts welcome.
Offline Kleinstein
Re: Data Precision 8200 - 100V range out of pot range during adjustment
« Reply #1 on: October 16, 2023, 02:47:17 pm »
There are relay contacts in series with the 10 K +90 K (incl. trimmer) gain setting resistors. Especially K2 in series to the 10 K may be an issue with relay contacts that were not used for a long time. A few cycles and whatching the change may be a good idea.
The following users thanked this post: Rax
« Reply #2 on: October 16, 2023, 09:04:28 pm »
My other thing to investigate is whether the 130V rails are in good shape. It's been a while since I last looked at this, but I think the Q3-Q7 is the stage doing the 10x and if the rails are too low maybe that robs the stage of enough headroom to go as high as it needs to.
Offline RandallMcRee
« Reply #3 on: October 16, 2023, 10:08:23 pm »
Your description makes it sound like that trim pot is busted. Easily happens.  Why don't you confirm that hunch, first?
The pot can only trim a small amount, though, so perhaps look at your working unit to confirm correct voltages around the trimmer and its associated resistor.
Offline doktor pyta
« Reply #4 on: October 17, 2023, 06:00:50 am »
Dirty relay contacts are common issue of DP8200.
Fortunately they are easy to open an clean with paper soaked with IPA.
Offline Testtech
« Reply #5 on: October 18, 2023, 09:18:54 pm »
There is a 68K resistor, R51, near the rear of the board that often burns up or changes value. Also the hv amp power transistors and their drivers can fail.