« on: September 27, 2018, 03:46:42 pm »
My Datron 4000A Calibrator is blowing tantalum capacitors again!
So I replaced all tantalums on PC Boards 400444 and 400445 with same uF tantalum caps, I upped the voltage with Kemet tantalum caps .
Now on the 400451 Power Supply board Datron used 22uF tantalum capacitors.
Yet if I read the datasheets on the 7800 and 7900 voltage regulators correctly, they suggest anything higher than .33uF.
( A 0.33 uF or larger tantalum,
mylar, or other capacitor having low internal impedance at
high frequencies should be chosen. The capacitor chosen
should have an equivalent series resistance of less than
0.7ohm. The bypass capacitor should be mounted with the
shortest possible leads directly across the regulators input
terminals. )
Is it possible to use different value uF caps in this location as I have a good supply of 10uF tantalums available?
« Last Edit: September 27, 2018, 07:32:34 pm by Johnny10 »
Logged
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B