Author Topic: Datron 4708 firmware need  (Read 2087 times)
Offline Cage03Topic starter
  • Country: cz
« on: April 09, 2020, 05:57:39 pm »
Hi, Have anybody Datron 4708 eprom dump file?
Online coromonadalix
Re: Datron 4708 firmware need
« Reply #1 on: April 09, 2020, 08:43:07 pm »
At KO4BB  website you have some manuals and some stuff,  but no eeprom dump
http://ftb.ko4bb.com/getsimple/index.php?id=manuals&dir=06_Misc_Test_Equipment/Datron
The following users thanked this post: Cage03
Offline doktor pyta
    • Vintage Test Gear
« Reply #2 on: April 10, 2020, 08:53:46 am »
If You  won't find anything within a week I will try to prepare dumps for You.
  • Country: cz
« Reply #3 on: April 10, 2020, 10:59:38 am »
I only found the dump file 4808. I would be very grateful for your help.
Offline doktor pyta
    • Vintage Test Gear
« Reply #4 on: April 11, 2020, 12:03:27 am »
OK, I'm faster than I thought.
Below You will find dump of M18 EPROM.
http://rfscientific.eu/sites/default/files/articleextra/eprom_datron_4708.zip
The following users thanked this post: coromonadalix, Cage03, the Chris
Offline doktor pyta
    • Vintage Test Gear
« Reply #5 on: April 14, 2020, 12:36:04 am »
@Cage03
Can You share some photos, details about Your repair?
Was Your M18 EPROM corrupted?
  • Country: cz
« Reply #6 on: April 14, 2020, 04:26:19 am »
Today I will continue to repair and give here some more info and photos.
  • Country: cz
« Reply #7 on: April 14, 2020, 03:28:37 pm »
My M18 eprom has a fault, 8 bytes are OK, another 8 bytes always have 0 on the second bit.
Other differences are due to different firmware versions, but 4708 is running.
There is still some defect on the AC board because the Reference Divider
control voltage is only from 0 to 8 V and when I remove the AC board is to 20 V.
Offline doktor pyta
    • Vintage Test Gear
« Reply #8 on: April 15, 2020, 09:52:56 am »
First of all check output of M3 optocoupler on REF Divider board.
It carries 16kHz reference clock for PLL in AC synthesizer.
When this signal is not present, the amplitude and frequency of AC are behaving very strange.
The following users thanked this post: Cage03, the Chris
Offline perdrix
« Reply #9 on: April 15, 2020, 06:12:13 pm »
IIRC there are two programmable devices on the 4808 CPU board.  I assume this is for the main ROM - M18?  Has anyone ever grabbed the content of the M3 address decode PAL?
Where did you find that ROM image please?
Thanks
