Author Topic: Fluke 199C broken! Help to repair  (Read 12124 times)

0 Members and 2 Guests are viewing this topic.

Offline NicolaFabbriTopic starter

  • Regular Contributor
  • *
  • Posts: 102
  • Country: it
Re: Fluke 199C broken! Help to repair
« Reply #200 on: June 14, 2026, 07:57:26 pm »
@squadchannel sorry if I come back... but after some month where oscilloscope was ok, I found that keboard was not responding except switching on-off or doing this and entering in DFU mode. So I decide to reflash firmware, and I come back on first page where oscilloscope ask me the language, but keyboard don't work! Nether I'm able to simulate key via serial line.
But waveform is running!

Strange...
Any idea?  :'(
 

Online squadchannel

  • Super Contributor
  • ***
  • Posts: 1184
  • Country: jp
  • deepl translate user
Re: Fluke 199C broken! Help to repair
« Reply #201 on: June 15, 2026, 08:36:41 am »
Hmm, the firmware is working, and I don't really have any ideas. :-//

@asis, what do you think?
 

Offline asis

  • Frequent Contributor
  • **
  • Posts: 364
  • Country: ru
Re: Fluke 199C broken! Help to repair
« Reply #202 on: June 15, 2026, 08:14:20 pm »
Hi,

First, I would carefully inspect the PCB for the condition of the tantalum ECAPs.
Then, using the scope's schematic, I would measure all supply voltages, paying attention to the presence of pulsation.
Then, I would disconnect the keyboard flex cable and, by trial and error, determine which control commands the KY scope responds to via the OC4USB cable.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf