Switches are clean, yes. To be sure i measured (with REL) every contact and its 0,04 or less Ohms. Same for the PCB mounting points - they fit, none is missing or bent. Resistance also less than 0,10 Ohms.
Not sure if i did, what you want, Harry but here are some results:
Fluke 8060A (which i can trust / its tested with a cal source) VS 8600A both in ACV mode
200 mV range
shorted input in 200 mV range:
8060A: 0,01 mV
8600A: 1,98 mV
200 mV range
100,00 mV sin 50 Hz input:
8060A: 100,00 mV
8600A: 100,05 mV
---
2 V range
shorted input in 2 V range
8060A: 0,0000 V
8600A: 0,0020 V
2 V range
1 V sin 50 Hz input:
8060A: 1,0000 V
8600A: 0,9970 V
I'm not sure what to make out of this, because any difference could also be due to difference in calibration, right?
/edit: I just noticed if i use "AC Volts" mode and e.g. 200 mV range, there is a voltage at the "input" present corresponding to the value shown by the meter. In addition, if i select "AC mA" mode and e.g. 200 µA range, i see 22 +/- mV between "V/Ohm in and LO" which is also shown in the display of the instrument ... as "mA"
So somehow a tiny voltage creeps all the way to the "input" jacks.
/edit2: With the AC Board removed, the "phantom voltage" is gone from all input jacks. Even if i remove the AC V IN Connection (a single wire connected to S2A (mA) Switch) these "ghostly low voltages" are present.