So I checked the pullup resistors they are basically almost spot on at 1.48K (both of them) however I'm measuring in circuit.
Iterrupts, I treid with the Saleae logic analizer, there's absolutely no traffic that I can see in the Option line, unless I am mistaken.
And also, I have to rectify a statement: it is true that there's always unreg power to the GPIB connectr, however the gnd ref is controlled by the main switch so the board does get powered on when the power switch is pressed. However, and I did test this, using one of the two GND lines and unreg +V, it is possible to keep something always on inside the meter - i.e. an OCXO osc, that then is always ready when the whole meter is turned on.
I am struggling with the brand new Siglent SDS1202X-E, which does have I2C decoding and a lot of other things but I have to learn that too! Uff. I could not see the signal levels yet. However, as I said, no traffic on OPTION interrupt.Surely I'm doing somethng wrong? It HAS to have some traffic.