Thank you both for your suggestions. I have not seen anything unusual in my visual inspections under microscope.
All relay solder joints are perfect and clean. No visible damage to the hybrid circuit board.
The hybrids used in my TDS744A are H2462G's:
https://w140.com/tekwiki/wiki/H2462The repair procedure followed was described here (thank you Terra):
In the above tekwiki link, there is a block diagram of the hybrid. I see no control line for 1M input impedance configuration. So, if not 50 Ohm, terminated in 1MOhm on the hybrid? The same tekwiki link has a detailed description of the M753 amplifier used on the hybrid board. The initial picture has a block diagram of the x1, x10 and x10 resistive dividers used. Those resistive dividers must be the 1M input terminators.
https://w140.com/tekwiki/wiki/M753#/media/File:1990-08-23_New_Attenuator_scheme_m753.jpgI have confirmed that the below are the DC offsets on Ch4 for the different gain settings. The x1, x10 and x100 are the different resistive divider inputs on the M753 amplifier.
x1 (10V/div, 5V/div, 2V/div) : 40mV offset, 700 KOhm input impedance
x10 (1V/div, 500mV/div, 200mV/div) = -240mV offset, 225 KOhm input impedance
x10 (100mV/div, 50mV/div, 20mV/div, 10mV/div, 5mV/div, 2mV/div and lower) = -13mV offset, 680 KOhm input impedance.
I will take the unit apart again to see if I can identify any resistive dividers on the hybrid board. Perhaps I can confirm the above resistances.
Thanks,
/John.