I was at our local electronics store picking up some miscellaneous parts - it's about the only one left which doesn't specialize in TVs, computers and knick-knacks. I was surprised to see that they had a shelf piled with old test gear; several expensive HP analog meters and some stuff which I couldn't identify. But there in front was a fairly clean looking HP 3490A. This is the unit which has a built-in logic test suite and is supposed to be fairly accurate within its 5.5-digit range. There's a nice article about it in the HP Journal for August 1972.
There was no way to plug it in and test it, so I hazarded a price and the owner seemed almost glad to get rid of it. Though when we looked inside (to make sure it hadn't been stripped) he said it was lucky he didn't realize there was so much gold in it or he probably would have salvaged it a couple of years ago.
Anyway, home it went. Powered up OK, and the PSU seems to be providing the correct voltages. The logic test runs to completion with all the correct displays, except for an added decimal point which appears 2 digits over from the "right" one during parts of the test. However, every other display in the test suite as well as all the measurement ranges yields the "-12000.2" overload display (max range is 120000 counts). I still haven't gotten a really good sense of how this beast is put together, but I suspect that since the flow is Signal conditioner -> DC Amp -> ADC -> Logic -> Display that the fault probably lies in the DC Amp or the ADC. I thought it might be a quick fix when I found a toasted resistor (A2R8) in the DC amp but it turned out to only be the guard ground connection. That, unfortunately, hints that the DC amp may have taken a jolt.
That's a problem, because the big hybrid, A2U2, is complete unobtainium - in fact, the service manual pretty much says "if it's bad, exchange the board for a new one."
Nevertheless, I'm hoping I can learn a few things about this unit even if it's not repairable without an expensive donor. There are plenty of other things it could be, but apparently either the DC amp is pinned to one rail or the ADC front end is likewise malfunctioning. The logic test, if I decipher the block diagrams correctly, does not test the signal path directly, only the digital portion; however, the other tests do attempt to verify the DC amp, ohms converter, and reference voltages. So I'm leaning toward that deceptively sparse DC amp board which sits between the signal converter and everything else.
Oh well - now that I have all my Fluke DMMs purring along, time for something a little more challenging, right?