Thanks for explaining the circuit theory, guys. I think this makes a little more sense now. Would I be correct in stating that the floating +S ground allows all of the functional blocks to shift their common references "automatically" as the output voltage changes? How does -S, the output "common," play into this? The operation is different than other control circuits I'm used to, where the reference points are fixed. What's the advantage of a floating design?
As for the troubleshooting, we are making some progress! I probed the test points in the Current Input Circuit while in CC mode and discovered that TP 14 (base of Q2A) was not behaving as described. It turned out that CR5 was indeed faulty (shorted) as xavier60 suggested. I presume that this short was causing Q2 to conduct constantly? I noticed that the OR gate now seems to work as expected, with the inputs at CR3 and CR4 switching according to CV or CC mode.
After fixing CR5, the current limit now reaches the full rated amount (1.8 A in over-range)
. However, I am still experiencing the weird 60 Hz modulation in CC mode
.
I took duak's advice and checked all of the voltage references again. I admit that I don't have the best probes for high sensitivity measurements, but I couldn't really detect any 60 Hz ripple on the rails. The 6.2, -6.2, and 12.4 V rails were all within 1 mVpp, too.
For confirmation, I also added the extra 490 µF cap across C10. There wasn't any change in the output ripple, although the ripple on the 12.4 V rail dropped from 4.5 mVpp to 1 mVpp. Either way, it's in spec, but at least I know my extra cap isn't toast. I also put the capacitor across the other filter caps, C12 and C14, and did not see any significant change on the output ripple. So, I don't think the ripple is a filtering issue.
So, thus far, I had only been able to observe 60 Hz ripple on the output. I couldn't find it anywhere else in the circuit. That may have changed now, but more probing is required--I found 60 Hz in the Voltage Clamp Circuit at TP 43. I think this circuit is only used in CC operation, so it makes sense that it would affect that operating state.
I have three waveforms below that show different states based on the position of the voltage control knob, relative to the CV/CC transition point, with the current limit set to 1 A. In full-on CV mode, the output ripple is normal and there's an obvious 60 Hz sine wave on TP43. At the transition point, the 60 Hz waveform changes, but it curiously corresponds to the 60 Hz output ripple (though inverted). In full-on CC mode, the 60 Hz waveform goes away entirely, and yet the output ripple remains.
I'm not sure what to make of this just yet, but I think I need to start probing some other areas to find where the 60 Hz interference is coming from. I actually disconnected the Voltage Clamp Circuit (via CR30 and CR32) in my testing a few days ago, just to rule out extra components, and the 60 Hz ripple was still there, so I suspect it might be coming from another place and just manifesting itself in the clamp circuit. I'm thinking the -4.3 V rail might be worth checking again now that the Current Input Circuit is fixed.