Author Topic: HP34401A  (Read 1991 times)

0 Members and 1 Guest are viewing this topic.

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
HP34401A
« on: February 04, 2024, 09:02:53 am »
Solid Beep on boot, I have a hp34401A and it won't stop beeping on boot the display switches ons but nothing works, the beeping does stop if I change to use the rear jacks but it still doesn't go past the boot screen
https://youtu.be/GEM9esejeUU
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #1 on: February 04, 2024, 09:04:40 am »
More photos
 

Offline iMo

  • Super Contributor
  • ***
  • Posts: 4801
  • Country: pm
  • It's important to try new things..
Re: HP34401A
« Reply #2 on: February 04, 2024, 09:47:49 am »
fyi
PS: do start with the checking of all supply voltages..
Download the Service manual, there is the schematics with all the components on the pcbs..
« Last Edit: February 04, 2024, 09:52:51 am by iMo »
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #3 on: February 04, 2024, 10:47:57 am »
5v was fine and 18v looked fine, but I will look to order a replacement for that part any ideas about the buzzer
 

Offline trobbins

  • Frequent Contributor
  • **
  • Posts: 767
  • Country: au
Re: HP34401A
« Reply #4 on: February 04, 2024, 11:23:41 am »
Suggest start with all the voltage rails - there are more than two rails, and they include regulator circuitry.  Afaik there is/are beep(s) as mains voltage increases (eg. when using a variac), and likely related to under-voltage lockout style activation of sections of circuitry.
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #5 on: February 04, 2024, 11:41:03 am »
Just noticed the 3amp fuse in the rear holder is burst just an fyi
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #6 on: February 04, 2024, 12:30:07 pm »
Any further steps for diagnosis or fixes
 

Offline trobbins

  • Frequent Contributor
  • **
  • Posts: 767
  • Country: au
Re: HP34401A
« Reply #7 on: February 04, 2024, 01:20:37 pm »
The fuse has blown, so do you have a Variac, or can power from lower than mains, or have a bulb limiter?  Start by disconnecting the 8-pin connector from the transformer. But you need to have the service manual, and need to appreciate the power supply section schematic first ?????
 

Offline factory

  • Super Contributor
  • ***
  • Posts: 2890
  • Country: gb
Re: HP34401A
« Reply #8 on: February 04, 2024, 01:35:49 pm »
The 3A fuse is for the current measurement function, I can't imagine it would stop the DMM from working on other functions?
Mains fuses are much lower values.

David
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #9 on: February 04, 2024, 02:01:33 pm »
Sadly I do not have either of those, Ive been looking at the schematics for the power supply and I really doubt its an issue with that
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #10 on: February 04, 2024, 02:13:48 pm »
Just noticed that this part is getting very hot but its still supplying a pretty clean 5v so Im a bit stumped
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #11 on: February 04, 2024, 02:24:48 pm »
Still Stumped on why when the front selector is pressed to use the rear inputs and the multimeter is rebooted the beeping stops
 

Offline iMo

  • Super Contributor
  • ***
  • Posts: 4801
  • Country: pm
  • It's important to try new things..
Re: HP34401A
« Reply #12 on: February 04, 2024, 02:54:27 pm »
Just noticed that this part is getting very hot but its still supplying a pretty clean 5v so Im a bit stumped
That 5V voltage regulator powers the outguard logic only. There are i/o interfaces (HPIB, serial) driven by an MCU where the chips might be damaged. Also the tantalum capacitors need to be checked.
The +/-18Volts in the inguard logic (with the ADC, input stuff, main CPU etc) go through various serial zeners (3.2V) thus creating +/-15V for all important analog circuitry there ( :palm: ). Doublecheck them as well, as they get damaged easily.
« Last Edit: February 04, 2024, 03:02:46 pm by iMo »
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #13 on: February 04, 2024, 02:57:55 pm »
Ok thanks, could that be the root cause?
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #14 on: February 04, 2024, 02:58:50 pm »
Is this meant to be unpopulated
 

Offline iMo

  • Super Contributor
  • ***
  • Posts: 4801
  • Country: pm
  • It's important to try new things..
Re: HP34401A
« Reply #15 on: February 04, 2024, 03:05:28 pm »
Yes, unpopulated, that goes to the RS232 connector, there are 2 free pins without a signal on them.
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #16 on: February 04, 2024, 03:36:53 pm »
I am assuming by serial zeners you mean this i checked them at it ends at 14.8v
 

Offline tggzzz

  • Super Contributor
  • ***
  • Posts: 19627
  • Country: gb
  • Numbers, not adjectives
    • Having fun doing more, with less
Re: HP34401A
« Reply #17 on: February 04, 2024, 03:44:07 pm »
I am assuming by serial zeners you mean this i checked them at it ends at 14.8v

To understand the input protection circuits, there's a good explanation in "The Art of Electronics x-Chapters". https://x.artofelectronics.net/
There are lies, damned lies, statistics - and ADC/DAC specs.
Glider pilot's aphorism: "there is no substitute for span". Retort: "There is a substitute: skill+imagination. But you can buy span".
Having fun doing more, with less
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #18 on: February 04, 2024, 03:49:32 pm »
Thanks will take a look at this tmr morning but Im assuming what I did was incorrect?
 

Offline iMo

  • Super Contributor
  • ***
  • Posts: 4801
  • Country: pm
  • It's important to try new things..
Re: HP34401A
« Reply #19 on: February 04, 2024, 03:55:04 pm »
Those are not the power supply zeners I mentioned above (those are around the ADC part, look at the schematics and pcb view in the Service manual).
CR304,305,306,307
CR400, 401
« Last Edit: February 04, 2024, 04:04:35 pm by iMo »
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #20 on: February 04, 2024, 04:04:06 pm »
Oh you are referring to the ones here correct?
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2898
  • Country: 00
Re: HP34401A
« Reply #21 on: February 04, 2024, 04:13:37 pm »
Did you get it like this or did it suddenly break? If so, did anything particular happen to it before it broke?

Did you measure both voltage and ripple on all power rails (+/- 18V floating, +5V floating, +5V earthed, 6VAC for front panel)? Ideally you use a scope to measure.

The blown fuse and damaged C109 cap (in parallel with the MOV) both suggest excessively high input signals, though on different inputs. But otherwise after the power rails I'd start looking at the floating or front panel logic circuits. Study what the service manuals says about start up steps (does all segments lighting up indicate correct operation of the front panel?), and based on that identify a suspect. Then look for clock, activity on address bus, reset line, etc.

Can you connect to it using RS-232 or GPIB? Maybe it's just the front panel that locks up.
« Last Edit: February 04, 2024, 04:17:03 pm by alm »
 

Online J-R

  • Frequent Contributor
  • **
  • Posts: 994
  • Country: us
Re: HP34401A
« Reply #22 on: February 04, 2024, 10:33:59 pm »
I say focus on testing all the capacitors right away. You might be able to do some in-circuit, but if you can't tell for sure, remove for proper testing.  If you search for 34401A repairs, you'll see a lot of folks just replace all of them, especially on the older units.  Maybe you will get lucky and quickly find one that is shorted.
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #23 on: February 05, 2024, 12:09:22 pm »
Ok all the caps are fine anything I shoud check next
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #24 on: February 05, 2024, 12:14:03 pm »
all the big ones
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #25 on: February 05, 2024, 02:07:42 pm »
Ive noticed that these ic and the MCU are sitting around the mid 40's is this to be expected?
 

Offline Paceguy

  • Regular Contributor
  • *
  • Posts: 178
  • Country: ca
Re: HP34401A
« Reply #26 on: February 05, 2024, 03:27:51 pm »
Open power supply filter caps can cause regulators to run very hot. Did you check the ESR of all the power supply caps ?
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #27 on: February 05, 2024, 04:38:44 pm »
I sadly don't have a esr meter but I have checked the resistance using a multimeter and they are all very high so I don't think thats the issue
 

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 14272
  • Country: de
Re: HP34401A
« Reply #28 on: February 05, 2024, 05:10:33 pm »
DC resistance is very different from ESR.  High ESR would cause increasing power supply ripple. Instead of looking at the ESR one could also check the supply ripple with a scope and as a last resort a DMM (in AC mode).

The CPU can get a little warm, as it is a relatively old type and still running on 5 V. So it will use some power. It would be a bad sign if it is not getting warm.
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5960
  • Country: ca
Re: HP34401A
« Reply #29 on: February 05, 2024, 06:15:15 pm »
i would say,  yep  could be a case of bad capacitors ...  and while they are removed,  you could replace the small tantalums near them too   

with hot air ???  not an iron ..
 

Online J-R

  • Frequent Contributor
  • **
  • Posts: 994
  • Country: us
Re: HP34401A
« Reply #30 on: February 06, 2024, 12:10:40 am »
 
The following users thanked this post: bingo600, coromonadalix

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #31 on: February 08, 2024, 03:57:44 pm »
Does anyone have a list of the sizes of the different tantulum caps
 

Online J-R

  • Frequent Contributor
  • **
  • Posts: 994
  • Country: us
Re: HP34401A
« Reply #32 on: February 09, 2024, 05:11:59 am »
Does anyone have a list of the sizes of the different tantulum caps
The link I posted has the service manual which has all that info.
 

Offline lucianTopic starter

  • Contributor
  • Posts: 25
  • Country: sg
Re: HP34401A
« Reply #33 on: February 09, 2024, 01:18:29 pm »
It doesn't have the specific sizings
 

Offline JohanH

  • Frequent Contributor
  • **
  • Posts: 627
  • Country: fi
Re: HP34401A
« Reply #34 on: February 09, 2024, 02:00:47 pm »
It doesn't have the specific sizings

Here's the list of tantalums that I wrote down once, from the manual, but please double check:

C552, C554, C753 (main PCB)       1 uF ±20% 35 V tant (+18V rail) B-case (3.5 x 2.8mm)       
C616, C617, C602, C603, C604 (display)      1 uF ±20% 35 V tant (+18V rail) B-case (3.5 x 2.8mm) 
C411   4.7 uF ±20% 10 V B-case (3.5 x 2.8mm)   (C411 wrongly marked as L411 on component placement diagram)   
C311, C312, C322, C323, C557, C559     22 uF ±20% 20 V D-case (7.3 x 4.3mm)
C706, C512    47 uF ±20% 10 V D-case (7.3 x 4.3mm)

 
The following users thanked this post: coromonadalix

Online J-R

  • Frequent Contributor
  • **
  • Posts: 994
  • Country: us
Re: HP34401A
« Reply #35 on: February 09, 2024, 11:53:29 pm »
It doesn't have the specific sizings
OK, I thought you meant values not dimensions.  The service manual does have part numbers listed which can direct you to that info and an appropriate replacement.