I picked up this K2400 as parts and repair on the ePay. The firmware was corrupt, which was an easy fix for the no-boot issue.
Now that I've had a chance to start function verification, I've found that it always has about -240 VDC present on the Input/Output jacks.
I can adjust the voltage and current, and it does output voltage and current, but not at the set values. It won't go through calibration as the outputs are too far off.
It doesn't appear that a schematic is available, though the service manual is and has some good information. I also see that others have been through this sort of issue, though I don't see anything that might indicate a repeatable solution for the issue.
Voltage validations on the digital board are done. I need to run the same on the analog board. I did a thermal image, and U660 is smoking hot. A couple of other chips are relatively warm: an OpAmp and a 74HC409.
I figured I would open a thread to document anything I discover and any possible solution I can find. And, if anyone happens to know a schematic or potential solutions, I'm all ears.