« on: October 24, 2020, 11:28:01 am »
I thought I might post a curious problem on a Kikusui PLZ603W I have on the bench.
It is a 603W unit that thinks it is a PLZ153W model.
The load presented at the terminals is about 4X what is on the display.
Tried recalibrating unit but I cannot select the 603W in the calibration mode only the 153W.
The Calibration data table from Operations Manual can match differences exactly as per 153 to 603.
My assessment would be the firmware eprom.
This is a unit with unknown history.
As I said, everything works it is simply presenting 603 loads while showing 153 display. i.e. when set for .34A the actual load is 1.26A.
While proceeding to order an eprom programmer and asking around for a download of 603 Firmware (Firmware not available from Company), I thought I would post the problem.
The Kikusui PLZ-3W Models look to all have the same CPU board.
The difference being the number of add on load modules.
How the CPU determines which Model number baffles me at the moment.
Talked to a company Rep but they say not much data available for older units.