Hi everyone, sorry I've kept you waiting, had a work filled week and just now got to take care of everything.
@shakalnokturn Pin values are way down to what it should be. They changed a tad after I replaced R4, which was all over the place in values. Even after being removed from the board, when measuring, it starts with like 30kOhm, then jumps to 150megOhm, slowly creeps towards 200meg and then my meter bottoms out.
Checked with a second meter, had the same, so R4 went to retirement.
Caps where all replaced.
original R4 replaced
PIN13 - 2,6mV - 175,8mV
PIN17 - 8,73V - 9,26V
PIN18 - 7,8V - 8,35V
According to schematics, PIN17 gets power directly from the rectifier through R2 and R3. Both Resistors are fine, and measuring from a rectifier pin to PIN17 gave me 39,9 kOhms, which is what it should (40 kOhms) be.
So my guess is the rectifier has gone belly up somehow? On 232V~ the rectifier puts out 312V-. If I remember correctly, It should be more in the ballpark of 325V-?
PS: I also added the zener.
@SiliconWizard
Thanks for pointing the group out. They have quite a collection of Schematics, files, manuals and a lot of knowledge.