A couple of weeks ago, I got a 15-20 year old Mettler Toledo AT200 Analytical Balance (0.0001 g resolution). It seemed to work well for a couple of weeks. While everything was fine, I did notice that the separate PSU got a little warm while connected to the power (IR thermometer read an idling temperature of 36-37 degrees C on top of the PSU metal cabinet). Then a few days ago, the balance suddenly went dead as the automatic sliding doors were closing.
The power plug (primary side of the toroidal transformer inside the PSU) shows an open circuit, and therefore I guess the transformer must be considered dead. There are 4 secondary output wire pairs from the transformer that seem ok. The PSU also has 4 different outputs in one custom and 10-pin plug. (Sec.1: 12.8VDC@55mA, Sec.2: 6.5VDC@350mA, Sec.3: 10.8VDC@65mA and Sec.4: 16.7VDC@55mA).
As I looked closer at the PSU and the PCBA inside, it looked to me that it should be possible to replace the broken toroidal transformer power with something else. As all 4 transformer outputs go into different bridge rectifiers, I figured I could even use some DC source at appropriate voltage level(s) without being concerned about polarity. Each of the 4 voltages are regulated by a separate LM317T (SOT-220).
I connected my (0-15VDC) desktop power supply to the 6.5VDC circuit (the circuit with a LM317T connected to a 240R and a 2k2 resistor). It worked perfectly and gave a steady 6.53VDC output when supplied with an input of at least ~9.7VDC. All the 4 circuits was tested the same way, and they all seemed to work perfectly.
The problem occured when I connected and powered all the 4 inputs at once. Then multiple outputs were low, reading 2-3 volts. I figured out that I could power 3 outputs at the same time, but when I add power to the LM317T labeled "IC2" (the 6.5V one that can handle the most current), the PSU fails. As the IC3 and IC4 circuits (the lower part of the schematic) does not connect to IC1 or IC2 circuits, the IC3 and IC4 traces and components can not be causing the problem.
I have tried my best to draw a schematic of the circuit, using EasyEDA. Please see the enclosed file. I have mostly copied the physical component layout of the PCBA for easy recognition. When "CONNECTOR1 - AC INPUTS" pins 3 and 4 are powered, the components connected to pin 1, "GL1 - BRIDGE" and PTC thermistor "R1 - MOV" gets way too hot. (Desktop power supply set to max 350 mA to avoid damage).
I cannot find any faulty components. I guess I could start to desolder and replace components, but then there is a risk that something may get broken. Will a trained professional, or other enthusiast, like many of the eevblog forum readers, be able to lead my attention to the component(s) that are most likely to be faulty? I would for sure appreciate any tips.
Best regards,
nixxon