I have bought an second-hand analog oscilloscope.
It works fine but some problem in CH1.
When I use 0.1V/Div or above to measure the calibration port (2V square wave with 1k Hz), the amplitude of the waveform will be decreasing from 2V to about 0.6V in 5 minutes. The X-axis measure is fine.
After it decreases, if I tune to 50mV/Div or below for few seconds and then repeat the measurement in 0.1V/Div, the amplitude will be back to normal (2V).
However, even I doing so, it will have some change in CH1 DC gain and I need to calibrate it by opening the oscilloscope.
Such problem doesn't show in CH2 but just CH1.
The analog oscilloscope is MATRIX MOS-640CH (40MHz).
Does anyone know what's the problem with my oscilloscope and how could I fix this problem? Thank you!