had a recent failure on a ssg3k model, tons of tons of screws to remove
yes an HMC1118 failed, i may have spiked it ... you reeally need to dig out all the parts in the signal pathways, find the datasheet(s) if possible, i had 3x or 4x signal attenuators on the path, but where fine

etc ... and the HMC1118 at the end / output was dead
followed the signal with a scope and minimum x10 probe to have an high impedance ... it was easy to kill the signal while probing
I was sad to see many output protection parts not installed ...