My old 465M has been acting up lately, just when I need it the most! The narrow horizontal sweep problem was traced to one electrolytic cap on the main (interface) board. It is C582, a 2.2uf 200V EC that was a bit bulged at the bottom and only meas. 0.04uf on the LCR meter. It caused the +95V line to sag to about +65. I was unable to get the horiz. trace to begin on the left side of the CRT. The right side looked fine. Replacing this cap fixed it!
Now I have an issue on the vert. side that affects both channels. With both inputs set to gnd., the most I can lower the horiz. sweep line, on both channels, using the up/down position control, is one division below the center. Very odd. Further, the amplitude of the built-in cal sqr. wave will increase as the position knob is turned CW. The bottom will remain fixed, while the top of the sqr wave will rise, then both will rise together until they're both off the screen.
I measured the V's from both the + vert to the - vert and from each to ground, while varying the position knobs. The Wht/Blu (white with blue stripe) wire, not sure if this is + or - vert, remains fixed at 32.1Vdc for all positions from FCCW to FCW of the vertical position controls. The Wht/brwn (white with brown stripe) V varies from 32.1Vdc to 5.8V from FCCW to FCW. Both V's ref'd to gnd. So one appears stuck to the +32 V rail. Output transistor short? Shorted cap?
Fortunately, the service manual has some voltages listed on the schematics. Unfortunately, I don't have another scope to view the waveforms.
Update: Anyone know how to remove the guards? from around the vert. output transistors? Never seen anything like that before. These trnstrs are soldered in with what appears to be a plastic disk keeping them captive inside a larger circular frame. I removed two machine screws & nuts above where the bracket makes a 90° bend, but no joy going further.
Update II: I had to loosen the delay line from its anchors to the PCB in order to lift it to gain access to the vert. output xnstrs. These both tested good. Didn't find any shorted or open parts around/under them. Decided to reinstall them and power up in order to check V's. Tek's troubleshooting guide recommends shorting the bases of common-mode ckts together to null any differences between them and isolate the problem to a particular stage.
In order to proceed, I had some of the schematics enlarged to make reading them easier as well as the part board locater. It's a bit tedious going from the schematic to the part list to find the reference XY coordinates, then to the diagram for every single part. But it would be a royal mess without it!
Finally, I've read a few posts from guys who used to work in Tek service centers and saw many vertical board problems in 465M's come through the door. Hopefully they'll chime in here and give me a hint.