Hi!
Late status update to this thread: The PSU voltages were all perfect. Everything works fine except Ch1 trace is not visible. I used it a couple of minutes, and then two RIFAs blew along with a 68 Ohm resistor. So I replaced those and cleaned up in the PSU, and now the scope is back in business, but still missing Ch1.
However, I did find out something about the missing Ch1 trace. As you probably know, the 2445A displays an amplitude value in the top left corner of the display. When using Ch1, the value was large and negative. On Ch2, the value doesn't change when the input attenuator is changed. This is as expected. However, on Ch1, the amplitude figure went larger and larger as the V/DIV value was increased. So I started to suspect the input attenuator.
It turns out that when using the 1MOhm DC input setting, the resistance to ground is larger than 10 MOhm (i.e. O.L. on my DMM). So something is definitely wrong there. On DC 50 Ohm input, the resistance to ground is approx 50 Ohm. Note that there is no trace displayed for Ch1, even at the DC 50 Ohm input setting.
But when I opened up the scope and put in a jumper from the input connection on U200 (which is on the working Ch2) to the input connetion on U100 on Ch1, and specified the same V/DIV settings for both channels, I did get a a Ch1 trace which was a copy of Ch2, when using the built-in calibrator signal. I could control the vertical position of the two traces independently using the vertical position knobs of Ch1 and Ch2.
So this means that there is no problem with the famous hybrids, but with the input. I guess either with the relays, the chip that controls the relays, or perhaps the pins that connect the input attenuatur for Ch1 to the main board?