Hello everyone i'm new to this forum and electronics in general. I have this Tektronix 7603 oscilloscope that stopped functioning after not being in use for couple of years.
Now the problem is that the horizontal sweep is no longer functional. So i opened up the service manual and started to debug. Now according manual, the horizontal sweep is generated by the time base and in my case its the 7B53A which i checked by probing the SAWTOOTH OUTPUT using a DMM. Now with the slowest time on the time base, i saw a voltage ramp going from 0v to ~5V and then back to 0V so with this I figured the sweep generator might be functional so i moved to the mainframe instead.
From here i took a closer look at the horizontal amplifier and probed the P501 header since this came from the time base. This time though i saw a voltage ramp going from 0V to ~1.7v and then back to 0. So i figured that the sweep gets passed from the time base to this circuitry and the problem is the horizontal amplifier(maybe) but i'm not sure if this voltage range is correct or not.

I must add that the 'beam find' button is also functional as it causes horizontal deflection.

Now there was the unlabeled chip that i had suspicion as its the horizontal switch but i dont know how i'd go about checking if this chip is functional or not.
So any help would greatly be appreciated.
Here is a video demonstrating this problem: