Following on from the success of the Tek 7623A repair I thought I'd have a look at this 7603. This one came with a 7D01 Logic Analyzer plugin installed.
This thing is super dirty and crusty in some places inside, so I think one of the first things I'll do is clean it up a bit. I checked the voltage rails and they were out by maybe 1% but trimming the -50V brought them back within range.
Although, it seems this scope does have some major issues. The trace is doing some very strange things with and without plugins installed. Turning the Intensity knob has no effect, but turning up the Readout Intensity knob causes a strange pattern I don't like the look of, if you know what causes it let me know.  ???
With no plugins:
With plugins from 7623A installed:
With the Readout Intensity turned up:
Board locations:
Power rail test point Locations:
A5 Vertical Amplifier:
A6 Horizontal Amplifier:
A8 Z-Axis Amplifier
A3 Readout Board
« Last Edit: July 24, 2021, 06:28:43 am by sean0118 »
Re: Tektronix 7603 Oscilloscope - Repair Log
« Reply #1 on: July 24, 2021, 09:25:54 am »
Okay, I made a little bit of progress. The vertical amplifier outputs are way off from what I would expect. It's not a square wave, and it's way too slow (period should be ~1ms). The period doesn't change when I change the time base div either.
But, turns out there's massive 100Hz ripple on the -15V! it's pretty much dropping out to 0V then going to -20V. The multimeter was measuring about 14.8V which I thought was good enough, just shows I should have connected a scope to the power rails earlier.
I suspect the main bulk storage capacitor for the -15V may have gone bad, which would explain why it's 100Hz (rectified from 50Hz). But I guess there could be a faulty component pulling down the -15V too.
Vert Amp Outputs:
-15V rail:
Offline wn1fju
« Reply #2 on: July 24, 2021, 11:42:17 am »
I've been burned many times by voltmeter readings when checking power supplies, since the voltmeter will just give you an average value.  I always look at the power rails with a scope even if the voltmeter readings seem perfect.  In fact, even if a piece appears to be working correctly, it's always a good idea to consult the oscilloscope.
The following users thanked this post: sean0118
« Reply #3 on: July 24, 2021, 12:55:51 pm »
This will be a fun one. Whole power supply probably needs recapping. Then there’s the game of hunt the tantalum capacitor to play when it still doesn’t work. This journey is well worth it though. Good luck.  :-+
There’s a good thread here regarding the power supply and using adapter boards for the capacitor https://www.eevblog.com/forum/testgear/tektronix-76037613-repairs/
« Last Edit: July 24, 2021, 12:57:23 pm by bd139 »
Offline David Hess
  • DavidH
« Reply #4 on: July 24, 2021, 04:17:48 pm »
My first suggestion was going to be massive ripple from the power supply because of worn out input capacitors; I have seen it before.
« Reply #5 on: July 25, 2021, 04:37:06 am »
Yeah that's what it really needs. I'm actually planning to auction these off for charity here on the eevblog though (they are from work so can't sell them). I don't want to spend much on them because I won't be able to recover the cost, but hopefully whoever wins the auction gives it the full re-cap it deserves. Thanks for linking the other thread, it's good to have it here as a reference. ;)
Yeah it makes sense. I'd like to know what causes that pattern when the Readout Intensity is turned up. At first I was worried it was a problem with the CRT itself, but it almost looks like transistor Vds / Ids curves rotated 90 degrees, might just be a coincidence though.
I replaced C811 which directly follows the rectifier on the -18V line. I only had a 10,000uF which is really only a temporary fix as C811 was 14,000uF. I'm not sure if its ripple current rating is adequate for the long term either. I also added a wire to join the two -18V planes on the PCB that were connected by the multiple pins of C811. Unfortunately I forgot to take a photo of this while the module was out.
The ripple has improved a lot, now only ~1V on the -15V rail, probably due to the replacement being too small. The CRT intensity still doesn't adjust and the trace is still off centre etc. But the Readout Intensity adjustment doesn't produce that strange pattern anymore.
But, the -15V is now at -20V, so I assume the excessive ripple killed the regulator. Hopefully the -20V hasn't damaged anything.
A10 Rectifier:
A12 Signals out:
« Last Edit: July 25, 2021, 04:47:37 am by sean0118 »
« Reply #6 on: July 25, 2021, 10:06:09 am »
Okay, I made some more progress.  ;)
The main pass transistor (Q958) for regulating the -15V tested okay. But the smaller one that drives it (Q956) did not, not a complete short, but didn't pass the 'diode test'. Vbe and Vbc were both about 0.47, but were 0.8V in reverse instead of open.
Q956 is a 2N5859, but I replaced it with a 2N1893 because it's all I had in that package. It's not a good replacement, 0.5A vs 2A max collector current. I won't be running it long until I get a proper replacement.
Replacing Q956 was enough to bring it to life. However, it seems the oscillator on the reference board is dead, even though it's set to 1kHz only 4V DC is coming out. The photo taken below is using the 1kHz ref from my Rigol. The horizontal scale seems a bit off, might be a calibration issue, or maybe I still have some voltage rail ripple.
« Last Edit: July 25, 2021, 10:19:01 am by sean0118 »
« Reply #7 on: July 25, 2021, 10:33:22 am »
That’s looking pretty hopeful  :-+
Offline harrison@cs.unc.edu
« Reply #8 on: August 11, 2021, 02:26:38 pm »
I had similar problems with one of my 7603's. No high voltage led me to the rectifier board. No +5 or +130. I tested all the filter caps and
found that 5 out of 7 were leaky or just open. I replaced all of them with newer versions that are 105c tolerant. I tried to buy low ESR, but
the proper values were back ordered. After replacing all the caps, the +130 was still missing. I had checked for the raw 130 from the xformer
and it was correct. Still no +130.  The bridge rectifier was testing good while in the circuit, but still no +130. The problem was that one of the
xformer taps was intermittent. i had to remove one of the replaced caps to get to that connection. I resoldered and had all the voltages
present at the regulator board.
I didn't have a trace because no plug ins were installed. I put in the 7B53A and powered up again. The magic smoke left the scope!!
Two tantalums on the vertical and one on the 7B53 gave up life. I replaced them all and a few others. Calibration did not suffer and it has
been a good scope for me. I have 3 more on the way to repair and hopefully sell later.
nChapelHeel
« Reply #9 on: August 22, 2021, 05:33:39 am »
Okay I narrowed down the horizontal issue a bit more, I think the problem is actually in the 7B50 horizontal module.
One thing that threw me off a bit is the 4V ref is 1kHz +/- 20%, so it's not good as a reference for the horizontal. So I used the 1kHz ref from my Rigol instead, which should be the most accurate source I have.
I went through the calibration procedure for the horizontal gain of the 7603 (as best I could without a calibration module). First I set the 4V ref to DC and calibrated that to my multimeter, I set it back to 1kHz and swapped the vertical and horizontal modules. This gives vertical traces spaced horizontally by the 4V ref, which allowed me to calibrate the horizontal gain independent of the horizontal module.
Except when I swap the vertical and horizontal modules back to their correct locations the horizontal scale is still incorrect. I put the modules into my Tek 7623A  and it's off by the same amount. I should be getting 10 divisions at 0.1ms/DIV but I'm only getting 8.4 divisions:
Calibrating the horizontal gain (4V 0.5V/DIV)
Calibrating the horizontal gain (4V 2V/DIV)
Tek 7603 1kHz 0.1ms/DIV
Tek 7623A 1kHz 0.1ms/DIV
Offline David Hess
  • DavidH
« Reply #10 on: August 23, 2021, 03:23:49 am »
Standardization is only required if you expect to swap modules between oscilloscopes which was never really intended.
Before I got a standardizer plug-in, I modified my 7A16A amplifier with a couple of test points at the interface so that I can connect a multimeter to measure the DC and AC signal voltage at the interface, which as I recall is 25 millivolts per division but someone should check me on that.  Many multimeters have plenty of resolution and accuracy at that level to calibrate the mainframe better than required.  AC calibration can be done with a low frequency square wave and either AC RMS or AC averaging with a correction factor of 11%.
« Reply #11 on: October 26, 2021, 03:36:39 am »
Okay it died again!  ::)
I left it on for a bit then came back to find the trace was gone. Measuring the power rails I found they were all much lower than usual, which pointed to an issue with the -50V rail as it's the reference for the others.
When turning the scope on I found the -50V rail increased very slowly and never reached -50V. I suspected the 5.6uF electrolytic capacitor C889 in the reference circuit had gone leaky, however I lifted one leg and found it to be okay.
I then found R890 (5.49k) had gone open. It biases the zener diode reference, so without it the zener voltage was much lower than normal and C889 charged very slowly.
I replaced it with a CPF15K4900FEB14 (5.49k, 1%, 25ppm, 1W), it's much smaller than the original resistor, however the service manual specs it as only needing 0.5W. I also re-soldered some of the regulator pins as they looked possibly cracked and some had way too much solder. One or two of the pads were already lifted slightly so it may have been touched up before.
With R890 replaced the scope works again and all the power rails are in spec.   :D
Offline youngwug
  • Country: kr
« Reply #12 on: October 28, 2021, 09:32:05 am »
What is the measured value after removing the R890(5.49k)?
« Reply #13 on: January 09, 2022, 09:13:28 am »
Hey sorry I just saw this, I think I measured that but I can't recall what the value was. I just measured the voltage over it as 40.39V (-49.32V to -8.93V), that's only 7.4mA / 0.3W so I think it's okay, I'm not sure what caused the original resistor to fail.
I think something else is wrong with it now though, the trace keeps fading out, disappearing completely at times. The trace intensity adjustment seems to have no effect. The display intensity for the 1ms label etc does work fine though...
Anyone have ideas what I should look at first?
edit: I forgot to mention, the power rails (-15V, +5V, +15V, +50V, +130V, -50V) look good, both DC and ripple. I'll have a look the the manual to check the CRT bias etc.
edit2: Now that it's had time to warm up the trace isn't disappearing but it's brightness is unstable and the trace intensity knob still has no effect. Seems to be a similar issue to this:
https://www.eevblog.com/forum/repair/tektronix-7313-trace-disapears/
« Last Edit: January 09, 2022, 09:46:12 am by sean0118 »
Offline David Hess
  • DavidH
« Reply #14 on: January 10, 2022, 02:38:08 am »
The DC restorer which levels shifts from low to high voltage in the z-axis circuit path could be leaking before failure.  You might inspect it in a dark room and see something.
« Reply #15 on: January 10, 2022, 05:50:28 am »
Thanks, that seems quite possible, is the DC restorer in the HV module?
Since posting I have probed the Z-Axis test point with a scope (see pic below), I was trying to follow the 'Adjust CRT Grid Bias' instruction in the manual (pg 3-6). It seems there's meant to be a voltage offset at that point which is varied by the trace intensity control, however I only measure a constant 0V. But, varying the readout intensity does produce voltage spikes at that test point.
I think this might point to an issue on the Z-axis amp board? As a fault in the HV module would load down the line and effect the readout intensity too?
I've also ruled out a few other things, the plug-in modules work in the other scope, the intensity trim pot seems to be good (resistance measured at end of cable) and the CRT grid bias adjustment still varies the intensity.
« Reply #16 on: January 10, 2022, 10:24:49 am »
I can't catch a break with this scope, I accidentally left it plugged in over night, but with the front switch off. Went to turn it on but the mains fuse was blown, put a new fuse in (bad idea I know) and it blew that one in a few seconds...   How is it possible for one scope to fail in so many ways?  :D
edit: One of the +18V rectifier filter capacitors is short, either C813 or C814
« Last Edit: January 10, 2022, 11:15:10 am by sean0118 »
« Reply #17 on: January 23, 2022, 10:18:36 am »
Okay, I got it powering up again, it was actually rectifier CR811 that was shorted, but I had already cut out C813 and C814 so replaced them too. They probably needed replacing anyway.
CR811 seems to be a D-72 package, I had some similar spares that were very close but 1mm too high. It gets sandwiched between the PCB and enclosure so the extra 1mm meant they wouldn't fit. I eventually found a suitable replacement from Vishay which is in the list below. I was under the threshold for free shipping so I ordered long life Kemet caps for C813 and C814, I also ordered a 15,000uF replacement for C811.
Vishay   VS-KBPC602PBF   200V   6A
KEMET   ALC70A153CD040   40V   15,000µF
KEMET   ALC80A103CD040   40V   10,000µF
SIBA   70-065-65/1.6A   Time Delay   1.6A
It powers up now but I still need to fix this z-axis / trace intensity issue...     ::)
« Last Edit: January 23, 2022, 10:33:10 am by sean0118 »
« Reply #18 on: January 23, 2022, 01:05:46 pm »
Okay sorry to bump the thread again but I think I found the cause of the z-axis issue...   U99C is dead!
It's part of a CA3046 transistor array, the other transistors in the array test okay (just measuring BE, BC diode drops with a multimeter).
The exact same IC failed in my 7623A, maybe it's a common issue?
https://www.eevblog.com/forum/repair/tektronix-7623a-oscilloscope-repair-log/
« Reply #19 on: January 25, 2022, 01:27:09 pm »
Okay, replacing U99C fixed the intensity issue, well it still seems a bit unstable but the intensity knob does work now.
But I left it running for an hour and something else has failed... this is fault #7, must be close to a record...
The readout display has now collapsed vertically, and it moves with the trace position (I don't think it's meant to do that?). If I feed in the 4V ref the trace is the correct amplitude but I can't get it to trigger.  >:(
No Input:
4V 1kHz in (1ms/Div):
Offline David Hess
  • DavidH
« Reply #20 on: January 25, 2022, 05:45:59 pm »
The readout, and other traces, will shift with trace position because of thermal balance issues but it should not follow it.
« Reply #21 on: January 26, 2022, 12:58:51 am »
I'm starting to think the fault might be in one of the plug in modules (7A18N, 7B50), I can't get it to trigger when installed in the 7623A either. The 7623A doesn't have a readout board installed so makes it difficult to troubleshoot.
I'm guessing the readout might depend on the trigger system somehow?
Offline David Hess
  • DavidH
« Reply #22 on: January 26, 2022, 02:18:58 am »
The readout has different modes selected by a switch on the readout board.  One of the modes is free running where a readout is displayed at the time and also multiplexed during sweeps.  There is another mode where the readout is only generated after a sweep, which works better for photography and when the readout multiplexing would be annoying.
« Last Edit: January 26, 2022, 02:23:19 am by David Hess »
« Reply #23 on: January 26, 2022, 02:26:34 am »
The readout has different modes selected by a switch on the readout board.  One of the modes is free running where a readout is displayed at all time.  There is another mode where the readout is only generated after a sweep, which works better for photography and when the readout multiplexing would be annoying.
Yes good point, I tested that 'gate triggered' mode and it does work in that the readout is only displayed at the end of the sweep (but the display of the readout is still collapsed vertically etc).
« Reply #24 on: January 26, 2022, 03:28:42 am »
Okay, I opened the 7B50 horizontal module, tested and re-seated some of the transistors in the triggering circuit (they were okay) put it back in and the triggering works again...  might have been a poor connection somewhere?
One thing I did notice is the power rails in the module are down a bit (+14.5, -14.5V, +49.7V, -49.9V), I'm not sure what they are meant to be, but there are series resistors on the input to the module so maybe it's normal.
The issue with the readout still remains so I'm starting to suspect the readout board again...
Offline David Hess
  • DavidH
« Reply #25 on: January 26, 2022, 03:48:49 pm »
The voltage levels are not critical as long as they are stable.
There are three signals which matter for this coming out of the readout board:
- X/Y shutdown which disables the X and Y inputs
- The readout vertical signal
- The readout horizontal signal
The X/Y shutdown goes first to the horizontal channel switch on the horizontal amplifier board, and then to the vertical interface circuit on the interface board where the plug-ins connect.  The readout vertical signal goes directly to the vertical CRT amplifier.
What is weird is that both are failing because they originate and terminate in different areas, and the X/Y shutdown apparently *is* working for the horizontal signal.  That implies two different failures which seems unlikely unless something on the readout board could cause both.
Tekwiki has a cleaned up PDF of just the schematics that makes following these signals easy.
« Reply #26 on: January 30, 2022, 06:52:38 am »
Thanks, I hadn't seen the X-Y inhibit coming from the readout board on the schematic. Just in case it's useful for anyone with similar issues it comes out of the U2126 timer along with the Z axis signals.
It turns out the plug (P38) for the X-Y inhibit input to the main interface board wasn't plugged in properly, must have come loose when I removed the rectifier module earlier. With this fixed the readout display doesn't move with the trace anymore, but the vertical of the readout is still collapsed.
I've been probing around the readout input into the vertical amp board and I'm not sure what is going on. I can see the signal at the inputs to R414 & R408 (see below) but their outputs are at constant -15V -8V. Both sides of R429 are at -15V DC. Is -15V the correct bias or should it be closer to -50V?
I've measured most of the resistors in circuit and they seem okay, PNP transistors Q407 & Q415 (2N2907A) also test okay with a multimeter, I substituted them for general purpose parts too but no change. Could it be the hybred IC U450? I can measure the diode drops in the transistors between pins 1 & 2 and 4 & 5.
I feel I must be missing something obvious, otherwise I might need to start removing components to test them...
R414:
R408:
edit: I found a schematic in the 7623 user manual which shows the DC levels, mine seem roughly correct. I'm measuring -14.4V instead of -14.1V on the input of J431, but I assume that's a schematic error? Otherwise Vbe of the transistor internal to U450 has to be 1V...
« Last Edit: January 30, 2022, 08:00:11 am by sean0118 »
Offline David Hess
  • DavidH
« Reply #27 on: January 30, 2022, 03:34:07 pm »
Does the vertical work at all?  Does the position control work on the trace as expected while the readout stays along the bottom?  If so, then the hybrid part is working fine.
« Reply #28 on: February 05, 2022, 07:16:48 am »
Yep the vertical for the trace does work fine, that's a good point. Looking at the vert amp circuit again some of those resistors are very small in value, so there probably isn't meant to be a large voltage signal there. I'm starting to think the problem is actually on the Readout board.
Probing around the Readout board I found the Y Deflection line going into pin 7 of U2284 Format gen to be only 4V. It seems the line is lowest at U2284, the other side of R2279 is close to 5V for example. With it removed the line is no longer pulled down.
Also the channel address lines (pins 8, 1, 15) are higher in voltage than expected, 2.5V with a multimeter and with a scope I can see they are getting pulled up close to 5V. The wavefroms on one of the schematics I have suggests they should be getting pulled up to around 3V, but there are 1k pullups to +5V so might be normal...
I'm starting to strongly suspect U2284, might also try replacing U2250 because it's just a 7493 counter...
« Last Edit: February 05, 2022, 07:23:07 am by sean0118 »
Offline David Hess
  • DavidH
« Reply #29 on: February 05, 2022, 08:45:07 am »
It would be too bad if U2284 failed but I remember that happening in the past.
Many of these boards used Texas Instruments edge wipe sockets which have poor reliability, so the socket may have failed.  Reseating suspected ICs should be done.
« Reply #30 on: February 05, 2022, 09:07:27 am »
Yeah you're right, from looking around it does seem the Readout ICs might be prone to early failure:
https://groups.io/g/TekScopes/topic/7659141#142266
There's replacements for U2284 on ebay for a reasonable price so I might just buy one to see if it fixes it. I'm tempted to try adding a second resistor to pull it up harder first, although I doubt it will work.
Thanks for the suggestion, I have reseated all the Readout ICs and transistors but didn't see any change.
Offline David Hess
  • DavidH
« Reply #31 on: February 06, 2022, 12:24:40 am »
The failures I have seen are consistent with contamination through the plastic package, which would make it a matter of age and the packaging.
« Reply #32 on: June 04, 2022, 10:43:33 am »
Okay, I finally got back around to this project.  :D
I bought a replacement 155-0020-00 for U2284 from ebay (seller: rcflying-hobby). It came in the original packaging which was pretty cool, but the foam had degraded and started sticking to the legs, but it cleaned up okay using some DeOxit.
I swapped it over with the defective one and the readout is working again, so does confirm U2284 was the issue. One interesting thing, the readout is now at the top of the screen, was at the bottom before. I don't think the readout vertical position can be adjusted that far? Maybe it's a different revision IC?
Everything else seems to be working okay, but I still need to cal the horizontal timebase module...
« Last Edit: June 04, 2022, 10:46:45 am by sean0118 »
Offline dave_k
« Reply #33 on: June 04, 2022, 10:54:12 am »
Great work on the repairs, keep up the good work!
Offline David Hess
  • DavidH
« Reply #34 on: June 04, 2022, 07:39:12 pm »
Each plug-in slot gets readout positions at the top and bottom, and can use one or the other or both.  Not all plug-in types are consistent; early timebases use different readout positions than later ones.
The original IC must have been broken in that respect unless something else is broken now.  I do not know which readout positions the 7B50 uses because I only have later timebase plug-ins.
« Reply #35 on: March 23, 2024, 05:31:31 am »
Okay turns out the 7B50 timebase module has a "SWP CAL" trimmer on the front panel, so the horizontal cal was actually super easy.  :D
It's a little bit out once you get around the 50/100ns range, just the first few div on the left of the screen, maybe some small non-linearity somewhere either in the scope or timebase module? But pretty good considering 50ns is the limit of the 7B50 module:
1ms:
1us:
100ns:
50ns:
Offline David Hess
  • DavidH
« Reply #36 on: March 23, 2024, 10:23:15 am »
How do you know that the horizontal non-linearity is not from the 7603?
« Reply #37 on: March 29, 2024, 05:54:11 am »
Actually yeah I think you might be right, I swapped the modules into the 7623A and it looks better (maybe still out a little bit?). Looking at it again I'm not sure if it's non-linear, might be able to cal it out easily.
7623A:
7603:
« Reply #38 on: March 29, 2024, 07:54:17 am »
Okay I went through the horizontal amplifier calibration steps, seems to pass the "low frequency linearity" check. Low frequency horizontal gain adjustment seemed okay too.
While looking at the high frequency timing cal instructions I realised the 7B50 horizontal module can actually go to 5ns because of the x10 magnifier mode (although my time mark generator seems to struggle below 10ns).
Interestingly the 10ns high frequency adjustment capacitor (C588) is already at a limit so can't be adjusted further. I could adjust the 5ns adjust capacitors (C566, C586) though, which did improve the 10ns timing (slightly out on right side of screen still), although this did not improve the 50ns timing.
It's a bit of a mystery to me where the problem is, but seems to be getting near the limits of the scope?
10ns (before cal):
10ns (after cal):
50ns (after cal):