I have 3 probes for repair.
One was not detected by the scope correctly (but did not show error). It turned out that a flat flex cable in the compensation box was torn.
Another one showed spikes only where the edges should be, which looked like capacitive coupling. It turned out to be a bad connection under the tip cap.
The third one has a broken tip cap that is loose. I made a working replacement prototype.
All 3 probes have deteriorating rubber covers. I attempted to make a replacement from TPU95A-HF. I struggled with printing parameters such as printing speed, temperature and retraction. Only lowering the temperature seemed to help, but still there are some defects. I had just unpacked a fresh roll of QIDI TPU95A-HF. It is dispensed from a dry box. And the calibration cube looks perfect.
I wanted to try putting one on a probe, but desoldering the cable from either end does not seem to be very easy, so I decided to ask the community first.
Did anybody remove the cable from the compensation box? Is it easier than from the probe head? I have seen discussions regarding taking the head apart, but I don't want to do that if that is avoidable. It seems risky.