... the boost module have fail and gave an 37v for the ut71e after that it just run but giving 0.00000 on the voltage result or any other measurement...
From your pictures, the display and main selector switch appear to be behaving correctly. That suggests that the display controller IC (U9 M430FE425) and the I/O processor (U2 M430F149) survived the supply voltage being raised to 37V. The DMM IC (
U1 ES51966F), however, may not have survived (it uses separate input pins for different types of measurement, but now all measurement inputs appear to read zero).
A good way to start your troubleshooting would be to check the supply voltage to U1. It should be 5.5V and can be read between V+ (pins 62 & 63) and V- (pins 57 & 58) on U1. If you're lucky, the pre-regulator transistor (Q12) or the voltage regulator IC (U7 7201U55) that supplies the 5.5V, might have died and protected U1.