....so last night I went back to troubleshooting the ADC. All voltages (reference and supplies) were perfect but I noticed U6 is saturated and there is almost 500mV signal on pins 14,15 of U5 while it must be zero. Took a short while to find out that R4 (yes a green 0.1% 25ppm TC resistor) was open circuit. I replaced it with a shitty resistor for now and, voilĂ
It is fixed! well almost.
All readback voltages and currents are correct (with some small errors and offsets) EXCEPT the +6V voltage reading. With regard to the +6V voltage readback I quickly found out that the optoisolator U15 is not working because the input LED was open, so I need to replace that.
I went about measuring some of the precision 0.1% resistors around the ADC and a couple of other places and some of them are way off like 1-2% so I am planning to replace those as well before doing a calibration.
I can use my 34401A for calibrating almost everything except the 5A for which I must use my Fluke 289. Do you think this would be an acceptable calibration?