Hello,
in the past I have asked some help with calculations in this forum, and I got very good and instructive answers. So let me try again!
I am trying to study (self teach myself) something about amplifiers.
Concretely I have an old HP 461A (1Khz to 150Mhz) linear amplifier I want to understand properly.
This is a +40db (quite flat, +/- 1 db across the spectrum) amplifier.
Regarding noise performance, the datasheet says:
EQUIVALENT WIDEBAND INPUT NOISE LEVEL: Less than 40 uV in 40 db position.
I have a spectrum analyzer to do (more or less) precise measurements.
I did the following:
- Shorted the input of the HP 461A amplifier.
- Connected the output of the HP 461A amplifier to the input of the SA
- Verified that I get a flat noise floor across 0-150Mhz
- Using RBW of 1KHz I measured the noise floor to be -94dBm.
- Considering the -40db amplification, the amplifier is amplifying an equivalent noise of -94-40=-134dBm (again, at 1Khz RBW)
- I can deduce the density D (i.e., noise at 1Hz RBW) by subtracting by 10*log_10(1000)=30. So D = -164 dBm/Hz
- I have verified that, from D, I can get correct predictions for the noise level read by the SW at different RBW. For example, at 1Mhz RBW I get -64dBm.
Ok, so it looks like the HP 461A amplifier has an equivalent input noise density of -164dBm/Hz.
Taking the bandwidth to be 150Mhz, I can calculate:
[/list]
-164dBm + 10log_10(150e6) = -164 + 80.1 = -82dBm
of equivalent wideband noise input at 150Mhz bandwidth.
I can convert the power from dBm to Watts using the equation
-82dBm = 10*log_10( P/ 0.001)
and obtain P=6.30* 10^12 Watts (
https://www.wolframalpha.com/input/?i=-82+%3D+10*log_10%28+p%2F+0.001%29%2C++v%5E2%2F50+%3D+p+)
Finally, I can convert this to RMS volts (considering a 50 Ohm input/output) using
(v*v)/50 = P
and I obtain (
Measured equivalent wideband (150Mhz) noise RMS voltage: v = 17uV
which is compatible with the datasheet (which says <40uV).
Are these calculations OK or am I making some mistakes?The reason of my doubt is that, using the same style of calculations, I get a noise density expressed in V/sqrt(Hz) of only 1.4nV (
https://www.wolframalpha.com/input/?i=-164+%3D+10*log_10%28+p%2F+0.001%29%2C++v%5E2%2F50+%3D+p+)
This seems too god to be true? I had a look at the datasheet of minicircuits amplifiers (made in 2020, the HP 461A was made >50years ago in 1964) and the specs for LNA amplifiers often have >14.nV/sqrt(Hz) noise density. Also the HP 461A has a measured SWR <1.1, much better than some new Minicircuits amps. Is it really possible that this old HP unit is so good?
If it is actually that good, and the calculations are OK, could I use this amplifier directly to amplify the signal received by an antenna? Ot it would not be appropriate for that task?
Thanks!