I see NRDGS is set for varies measurements listed here, but I am not so sure whether they are rightly or wrongly used in the noise calculation. Just to continue my last post, here is my understanding:
In essence, NRDGS provides an alternative to the "TRIG" command to trigger the readings (sampling). It is a sort of "burst" mode. But, regardless what NRDGS is set to, how a reading is done is always defined by "NPLC". In other words, if each reading is a data point in the statistics, NRDGS setting is irrelevant.
To compare the noise figures between 2 machines, the (total) integration time per data point (not necessarily per reading/sample) for the calculation should be the same. To a good approximation the total integration time per data_point should be
"total_integration_time_per_data_point = NPLC x samples_per_data_point", assuming the value of a data point is taken as the average over samples per data_point.
Then, It will be interest to know, but, I did not test, as to what will be the actual difference on a 3457a between
data_point= "NPLC=10 and average over 10 samples"
and
data_point= "NPLC=100 and 1 sample"