A couple of years ago, I bought an used/abused Fluke 77 III. When I initially checked it out, everything seemed to work fine.
A few months later, I measured a 100 ohm resistor and discovered it was around 98 ohms (out of spec on the 320 ohm range). I then discovered that the 32M range was out of spec as well. Since I have never used this meter on any high energy or mains circuitry, I can only conclude that the previous owner damaged the meter and did not disclose it on the ebay listing.
After some testing, I discovered that the PTC was measuring 100k ohm instead of the 1k ohm. I replaced the PTC with a 1k ohm carbon film resistor that measured 970.8 ohm, as a test, and now all ranges are within spec.
What I'm hinting at is that whether the PTC is 1.8 or 1.71 or 2k ohm, it is likely not to affect the resistance measurements that much. In my example above, even though the PTC was 100x times more than its rated value, the 320 and 32M ohm measurements were not that far off and all the other ranges in between (3.2k, 32k, 320k, 3.2M) were fine.