Author Topic: Adret CV102 Voltage Source  (Read 4608 times)
Offline giovanniratTopic starter
  • Country: at
« on: February 25, 2017, 08:18:19 pm »
I've got an old Adret CV102 Voltage Source.
It works so far, but needs some cleaning and calibration.
Unfortunately I didn't found any documentation (User Manual, Service Manual,..) on the net.
I would really appreciate, if someone could provide such a documentation or any additional information regarding this Voltage Source.
Thanks
Offline bingo600
Re: Adret CV102 Voltage Source
« Reply #1 on: February 25, 2017, 09:33:06 pm »
Hope your french is ok
Remove the last .zip extension , it's a rar file.
But forum don't like rar
/Bingo
The following users thanked this post: Falkra
  • Country: at
« Reply #2 on: February 25, 2017, 10:10:29 pm »
Thank you very much bingo600  :-+
This document is exactly what I was looking for....
Great service
Offline Joel_S
« Reply #3 on: June 09, 2017, 03:10:35 pm »
Folks,
Would anyone have schematics of the Adret CV102 ?
Thanks to all !
Joel
Online Vgkid
« Reply #4 on: June 09, 2017, 03:18:40 pm »
Folks,
Thanks to all !
Joel
It is in the above 2 files. Going from memory, the second one.
Offline bingo600
« Reply #5 on: November 17, 2023, 08:07:21 pm »
UG as pdf
Offline Tillk
« Reply #6 on: November 25, 2023, 07:54:45 pm »
Hello,
I'll ask here again, maybe there is a new version.
The manual for the CV102 that you now find in EEVblog ends with chapter 6.
As you can see, the circuit diagram starts with chapter 7. Is anyone in
possession of this circuit diagram and can send it to me by e-mail?
My device has a fault and I want to try to repair it.
The voltage at the output is always 3.19V in the 10V range and 31.9V in the
100V range and cannot be changed.
Who can help me?
Best regards, T.K.
(Translated with www.DeepL.com/Translator, I speak German)
Offline Ankorjd
« Reply #7 on: December 11, 2023, 09:07:45 am »
Here are the schematics (Chapter 7).
I whish you success in your repair.
Jacques
Offline Tillk
« Reply #8 on: January 05, 2024, 08:44:27 pm »
Dear Mr. Jacques, thank you very much for the documents.
Now I can finally start looking for the error.
Best regards from Germany, T.K.
Offline Tillk
« Reply #9 on: January 16, 2024, 09:40:13 pm »
Hello to all, especially to Mr. Jacques,
the fault has been found, a defective tantalum capacitor 47µF on module 0131. now all capacitors have been replaced and the device is working again, but there is still a problem. the manual describes the adjustment. but the adjustment to zero with R22 does not work. i cannot achieve a positive value, the setting ends at -0.074 mV and never becomes positive. i have also measured that the
system clock operates at 1.624 MHz. how exactly does this value have to be correct? does the clock have to be set exactly again? the manual gives no indication. what else i have done...all TTL-Ics have been given a 47nF capacitor on the voltage pins for better performance.
Thanks and best regards, T.K.