and by the way no Factory Cal appears in the menus. Still the only cal I have in service menu is the user cal.
It's just a menu naming. Sure it's named "User Cal" because it's done by the user. But the purpose and result is the same as the factory cal (different place) - as is named in the log file and the TTL console.
The only difference from the "same thing" done in the factory - is that the factory actually verifies the instrument against a standard and prints a CERTIFICATE. They should have named the menu simply 'Calibration' (user or factory or whoever ... same thing).
callog.txt - you can erase that - nothing happens.
sernum.dat - serial number is NOT in /Secure directory (not as a file anyway) - I had corrupt /Secure folder - and sill had a serial number / MAC
All the calibration data is in framecal.dat.
Keysight recommends to perform user-calibration:
-
Every two years or after 4000 hours of operation
- If the ambient temperature is >10° C from the calibration temperature
- If you want to
maximize the measurement accuracy Performing User Cal will
invalidate your Certificate of Calibration. (InfiniiVision 2000/3000 X-Series Oscilloscopes Service Guide - Page 60)