I finally got the god damn thing fixed
Here is what you should do
1-install the Anritsu Master Software Tool (MST)
2-go to Product Update and download (retrieve) the firmware file inside the MST so it is ready
there you will see a button "Emergency Repair", click on that and it will explain what to do but I will tell you
3-connect the instrument to ethernet and turn it on in Recovery mode (hold softkeys #1-3-5 and power on)
when you are in the Bootstrap mode, choose Update OS using Ethernet and it will change to a blank screen with
one red statment saying it is waiting for Ethernet downloading connection...
4- in the MST enter the IP address of the instrument and click on Install New OS Now. and then just sit back and DO NOT TOUCH anything
no matter how everything seems weird or dead. (the message on the instrument screen will not change until much later) but it is actually
accepting the new OS from your computer
5-after quite a while the instrument will flash a new message saying the transfer is complete and it starts programming the ROMs on the instrument with new OS and it will finally go back to the first bootstrap screen
6- reboot the system and enjoy
However, the only problem that I have right now is that the tracking generator option is nowhere to be seen
my instrument passes all self tests and works perfectly but the tracking gen is not anywhere...
hope you can find a way to bring that back