I have a Beckman 3020 DMM that's reading about 1 ohm high. I've tested different probes, cleaned the jacks with Deoxit, checked the precision resistors inside (they're within spec), and next plan to carry out the calibration procedure (feed 190mV DC and adjust the rheostat to match while set to the 200mV range). What else am I not thinking of?