Here is a procedure for how to test and classify VNA cables - from Dr Joel Dunsmore.
Step 1: Put a load on the cable, put data->mem, then Data-Mem (NOT Data/mem), display LogMag. A perfect cable will be like -80 dB. Then bend the cable. Note the hightest peak, this will be the return loss or directivity stability of the cable. If there is a bad spot in the cable, that causes the match to change with bending, it will become apparent.
A very good cable will be below -50 dB, a cable worse than -30 dB should not be used.
Step 2: Put a short on the cable, put data->mem, then Data-Mem (NOT Data/mem) display LogMag. A perfect cable will be like -80 dB. Then bend the cable. Note the hightest peak, this will be the source match stability, and an indication of the phase stability. Because the short causes a large reflection, which we subtract with data-mem, any change in the cable loss or cable phase will show up as a big return loss (in dB, you are seeing 20 log (ref-test)). Thus, you are taking the difference of big numbers and you will see any defect in either phase, or return loss.
A good cable will be similar to above (-50) but because this tests both phase stability and reflection stability at the same time, it will show more errors and typically be worse than above.
Step 3: same as 2, but put data/mem and display phase. Bend the cable and see the phase stability. This will be the reflection tracking stability, and will be 2x the transmission stability. If you display logmag, this will be the magnitude stability of the reflection and transmission.
https://community.keysight.com/message/2595