Interestingly,despite all the hysteria posted about this possibility on this & other sites,the meter fared quite well.
It didn't "explode like a grenade",in fact the case plus sleeve contained everything well.
And this event wasn't because of the thing failing when used correctly.
It was hung across a 600V line,whilst still set to a current range!
When the only "CAT" rating was how many mice the Station cat could catch,Radio transmitter Techs used
AVO "8s" & the like to read high voltages all the time.
An AVO 8 on its 2500V range could sit there & read your wimpy 600V line all day & not flinch.
The three main things to remember when measuring high energy supply voltages are:
(1)
Check that the meter is configured to read volts (test leads in the correct sockets,etc) ,& is on the appropriate voltage range!(1a)
Check that the meter is configured to read volts (test leads in the correct sockets,etc) ,& is on the appropriate voltage range!(1b)
Check that the meter is configured to read volts (test leads in the correct sockets,etc) ,& is on the appropriate voltage range!OCD has a bad press,but it is just good practice in these circumstances.
(2) If it is at all possible,clip the probes in place,so you don't have to hold them--even better,if you can position them with the circuit deactivated.
(3) Unless you absolutely have to,don't hold the meter,place it on the ground,or some other convenient surface---the AVOs were so damn heavy,we were usually happy to do this!