The frequency range of interest for the interpolation test is about 40%...49% of the sample rate (and the way it looks, 30%...49% for the Rigol).
There is no need to sweep. Covering this range with 5 or 10 steps is certainly sufficient. Better try to get a stable view at each frequency. If you get AM, but can't get a stable trigger on the AM envelope, then better capture a single shot at each frequency step.
The timebase should not be faster than about 10/sample_rate (e.g. 5ns/div for 2GSa/s) in order that at least 1 or 2 periods of the AM envelope fit into the screen widht when the signal frequency is 49% of the sample rate. At lower signal frequencies, the envelope period will be shorter.