Author Topic: Difference Between Simpson 260-9SP & the 260-9S  (Read 1493 times)
Offline SG-1Topic starter
« on: May 02, 2018, 04:31:27 am »
Does anybody know what the difference is between the Simpson 260-9SP & the 260-9S ?
The instruction manual is the same for both instruments & the S seems to stand for the electronic protection circuit for the movement. There is a about a $70 difference in the price.
Section 1.2 of the manual covers the protection circuit. https://assets.tequipment.net/assets/1/26/Documents/Simpson/260_9SP/260_9sp_doc_2.pdf
Advice is a dangerous gift, even from the wise to the wise.
Offline BravoV
  • +++ ATH1
Re: Difference Between Simpson 260-9SP & the 260-9S
« Reply #1 on: May 02, 2018, 04:33:08 am »
Have you check here ? -> http://simpson260.com/
« Last Edit: May 02, 2018, 04:37:28 am by BravoV »
The following users thanked this post: SG-1
Offline Gregg
« Reply #2 on: May 02, 2018, 04:57:49 am »
I believe the Simpson P indicated a quick trip re-settable breaker; a small extra button on the front panel.
Offline kj7e
  • Damon Stewart
« Reply #3 on: May 02, 2018, 05:23:37 am »
The manual posted by the OP has a typo,"1.2 Overload Protection (9S Version)", should read "1.2 Overload Protection (9SP Version)".
The 'S' means Cat 3 to 600v, the 'P' means overload protected.