Author Topic: Error code -113 Undefined SCPI header  (Read 2594 times)
Offline magamaTopic starter
  • Country: pt
« on: July 24, 2025, 04:56:13 pm »
Hello everyone,
I'm new here to the forum and I'm developing a master's project that involves automating a collaborative robot to perform measurements on PCBAs. Currently, I'm testing the communication between Python, ESP32, the Cobot, and a KEITHLEY DMM6500 multimeter.
My problem lies in the communication with the multimeter. I can establish communication, and the multimeter performs the resistance reading, but when Python tries to read the value, an error appears on the multimeter: "error -113 undefined SCPI header, the SCPI command is not correct. verify that you are using the SCPI command Set"
I can't figure out what's wrong. Could you please help me? Below is my code:"import socket
import json
import time
import pyvisa
import csv
# ------------------------ MULTIMETER (Keithley DMM6500) ------------------------
# Initialize PyVISA Resource Manager
rm = pyvisa.ResourceManager()
multimeter_ip = "10.10.194.26"
# Open connection to the multimeter via TCP/IP
multimeter = rm.open_resource(f"TCPIP0::{multimeter_ip}::inst0::INSTR")
multimeter.timeout = 5000 # Set timeout for multimeter operations to 5 seconds
def configureResistanceMode():
    """Configures the multimeter for continuous resistance measurement."""
    try:
        multimeter.write("*RST") # Reset the instrument to its default settings
        multimeter.write("*CLS") # Clear the event registers and error queue
        multimeter.write("SENS:FUNC 'RES'") # Set the measurement function to Resistance
        multimeter.write("SENS:RES:RANG:AUTO ON") # Enable auto-ranging for resistance measurements
        multimeter.write("SENS:RES:NPLC 1") # Set integration time (Number of Power Line Cycles) to 1 for speed
        multimeter.write("TRIG:SOUR IMM") # Set trigger source to immediate (starts measurement as soon as possible)
        multimeter.write("SAMP:COUN 1") # Set sample count per trigger to 1
        multimeter.write("TRIG:COUN INF") # Set trigger count to infinite (continuous triggering)
        multimeter.write("INIT:CONT ON") # Enable continuous initiation (continuously acquires data)
        print("[Multimeter] Continuous resistance mode configured.")
    except Exception as e:
        print(f"[ERROR] configuring resistance mode: {e}")
def readResistancesDuring(seconds=3):
    """Reads resistance values from the multimeter for a specified duration."""
    readings = []
    end_time = time.time() + seconds
    try:
        while time.time() < end_time:
            # Use FETC? to fetch readings from the buffer when INIT:CONT ON is active.
            # READ? is often used for single measurements or when not in continuous mode.
            reading_str = multimeter.query("FETC?")
            value = float(reading_str.strip())
            print(f"[Multimeter] {value} Ohms")
            readings.append(value)
            time.sleep(0.5)  # Adjust as needed to control polling rate
        print(f"[ERROR] during continuous reading: {e}")
    return readings
def stopContinuousMode():
    """Stops the continuous measurement mode on the multimeter."""
    try:
        multimeter.write("INIT:CONT OFF") # Disable continuous initiation
        multimeter.write("ABORt")         # Abort any pending operations
        print("[Multimeter] Continuous mode turned off.")
        print(f"[ERROR] stopping continuous mode: {e}")
# ------------------------ ROBOT ------------------------
def connectETController(ip, port=8055):
    """Connects to the ET Robot controller."""
    sock = socket.socket(socket.AF_INET, socket.SOCK_STREAM)
    try:
        sock.connect((ip, port))
        return (True, sock)
    except Exception:
        sock.close()
        return (False,)
def disconnectETController(sock):
    """Disconnects from the ET Robot controller."""
    if sock:
        sock.close()
def sendCMD(sock, cmd, params=None, id=1):
    """Sends a JSON-RPC command to the ET Robot controller."""
    if not params:
        params = []
    else:
        params = json.dumps(params)
    sendStr = f'{{"method":"{cmd}","params":{params},"jsonrpc":"2.0","id":{id}}}\n'
    try:
        sock.sendall(bytes(sendStr, "utf-8"))
        ret = sock.recv(1024)
        jdata = json.loads(ret.decode("utf-8"))
        if "result" in jdata:
            return (True, json.loads(jdata["result"]), jdata["id"])
        elif "error" in jdata:
            return (False, jdata["error"], jdata["id"])
        else:
            return (False, None, None)
    except Exception:
def moveToPoint(sock, position, speed):
    """Moves the robot to a specified joint position at a given speed."""
    suc, result, id = sendCMD(sock, "moveByJoint", {"targetPos": position, "speed": speed})
    print(f"[ROBOT] Movement result: {result}")
    if not suc:
        print("[ROBOT] Error sending movement command.")
        return
    while True:
        suc, state, _ = sendCMD(sock, "getRobotState")
        if state == 0: # State 0 typically means robot is idle/stopped
            break
        time.sleep(0.2) # Poll robot state every 0.2 seconds
# ------------------------ ESP32 ------------------------
def sendCommandESP32(command, esp32_ip, port=3333, wait_time=1):
    """Sends a command to the ESP32 and waits for a response."""
    try:
        with socket.socket(socket.AF_INET, socket.SOCK_STREAM) as s:
            s.connect((esp32_ip, port))
            s.sendall((command + "\n").encode())
            print(f"[ESP32] Command '{command}' sent.")
            time.sleep(wait_time) # Wait for ESP32 to process and respond
            response = s.recv(1024)
            print(f"[ESP32] Response: {response.decode().strip()}")
        print(f"[ESP32] Error communicating: {e}")
# ------------------------ MAIN EXECUTION ------------------------
if __name__ == "__main__":
    robot_ip = "10.10.8.182"
    esp32_ip = "10.10.5.113"
    # Define robot joint positions
    point_home = [0.000, -90.000, 0.000, -90.000, 90.000, 0.000]
    point_igbt1_approach = [-35.103, -110.685, 108.715, -59.228, 72.368, -100.673]
    point_igbt1_measurement = [-35.103, -110.713, 111.545, -62.029, 72.368, -100.673]
    # Connect to the robot
    conSuc, sock = connectETController(robot_ip)
    print("[ROBOT] Connection successful:", conSuc)
    if conSuc:
        # Open CSV file to save measurement results
        with open("igbt1_measurements.csv", "w", newline="") as csvfile:
            writer = csv.writer(csvfile)
            writer.writerow(["IGBT", "Measurement Type", "Value (Ohms)"])
            # Check robot state and clear alarm if necessary
            if state == 4: # State 4 typically indicates an alarm
                sendCMD(sock, "clearAlarm")
                time.sleep(2)
            try:
                # Robot movement sequence for IGBT1
                moveToPoint(sock, point_home, 30)
                moveToPoint(sock, point_igbt1_approach, 15)
                moveToPoint(sock, point_igbt1_measurement, 1)
                # Configure multimeter for resistance measurement
                # Measurement 1 (R3 + R6) - Assuming this corresponds to a specific test
                sendCommandESP32("IGBT1_M1", esp32_ip)
                # Read resistances for 5 seconds
                r1_readings = readResistancesDuring(seconds=5)
                sendCommandESP32("STOP", esp32_ip) # Send STOP to ESP32 after measurement
                stopContinuousMode() # Stop continuous mode on multimeter
                writer.writerow(["IGBT1", "G-E (R3+R6)", r1_readings[-1] if r1_readings else None])
                # Measurement 2 (R1 + R5) - Assuming this corresponds to a specific test
                configureResistanceMode() # Reconfigure multimeter for next measurement
                sendCommandESP32("IGBT1_M2", esp32_ip)
                r2_readings = readResistencesDuring(seconds=5)
                writer.writerow(["IGBT1", "C-E (R1+R5)", r2_readings[-1] if r2_readings else None])
                # Measurement 3 (R4 + R6) - Assuming this corresponds to a specific test
                sendCommandESP32("IGBT1_M3", esp32_ip)
                r3_readings = readResistencesDuring(seconds=5)
                writer.writerow(["IGBT1", "G-C (R4+R6)", r3_readings[-1] if r3_readings else None])
                # Return robot to approach position and then home
                sendCommandESP32("STOP", esp32_ip) # Ensure ESP32 is stopped before moving robot
            except KeyboardInterrupt:
                print("Movement interrupted by user.")
                # Ensure all systems are stopped on interruption
                moveToPoint(sock, point_home, 30) # Move robot to home position
                print(f"[MAIN ERROR] An unexpected error occurred: {e}")
                # Ensure all systems are stopped on error
            finally:
                # Ensure all systems are stopped and connections closed
                sendCommandESP32("STOP", esp32_ip) # Final stop command to ESP32
                stopContinuousMode() # Final stop command to multimeter
                disconnectETController(sock) # Disconnect robot"
Offline KungFuJosh
  • TEAS is real.
Re: Error code -113 Undefined SCPI header
« Reply #1 on: July 24, 2025, 05:07:56 pm »
Hello everyone,
Welcome to the site.  I can't comment on your coding, but that error means the SCPI command is either unrecognized or incorrectly entered/communicated.
Have you tested the SCPI command(s) outside of your scripting?
Thanks,
Josh
Best Continuity Tester Ever
Best Continuity Tester Ever...Got Betterer!
The following users thanked this post: magama
Offline Rubo
« Reply #2 on: July 24, 2025, 06:02:05 pm »
You need to isolate which command generates that error.
I would issue every command in the chain one by one either with an if - else or a while loop and catch any errors.
That should narrow it down.
Offline gamalot
  • Correct my English
    • Youtube
« Reply #3 on: July 24, 2025, 06:08:32 pm »
You just need to try those SCPI commands one by one to find the problem.
Also, you should learn how to post code in the forum.
import socket
import json
import time
import pyvisa
import csv
    try:
    readings = []
    try:
    return readings
    try:
    try:
        return (True, sock)
    except Exception:
        sock.close()
        return (False,)
    if sock:
        sock.close()
    if not params:
        params = []
    else:
    try:
        else:
    except Exception:
    if not suc:
        return
    while True:
            break
    try:
    if conSuc:
                time.sleep(2)
            try:
            finally:
I'm a poet, I didn't even know it. |  https://youtube.com/@gamalot | https://github.com/gamalot
  • Country: pt
« Reply #4 on: July 25, 2025, 03:03:13 pm »
Thanks for the answer,
The resistance reading appears in the multimeter and the values that the multimeter is reading also appear in the VScode terminal.
The error message -133 appears on the multimeter screen.
The error only appears since I added "def lerResistenciasDurante(seconds=3):"
I made a mistake when publishing the code. The code that caused the error is:
def configurarModoResistencia():
    try:
        multimetro.write("*RST")
        multimetro.write("*CLS")
        multimetro.write("SENS:FUNC 'RES'")
        multimetro.write("SENS:RES:RANG:AUTO ON")
        multimetro.write("SENS:RES:NPLC 1")
        multimetro.write("TRIG:SOUR IMM")
        multimetro.write("SAMP:COUN 1")
        multimetro.write("TRIG:COUN INF")
        multimetro.write("INIT:CONT ON")
        print("[Multímetro] Modo resistência contínua configurado.")
        print(f"[ERRO] ao configurar resistência: {e}")
def lerResistenciasDurante(segundos=3):
    leituras = []
    fim = time.time() + segundos
    try:
        while time.time() < fim:
            leitura = multimetro.query("READ?")
            valor = float(leitura.strip())
            print(f"[Multímetro] {valor} Ohms")
            leituras.append(valor)
            time.sleep(0.5)
        print(f"[ERRO] durante leitura contínua: {e}")
    return leituras
« Last Edit: July 25, 2025, 03:19:58 pm by magama »
Offline TizianoHV
    • My Website
« Reply #5 on: July 25, 2025, 03:18:02 pm »
When you use "READ?"  the multimeter should be in "single shot" mode with: trig count = 1 (not INFinite) and init:cont off
    try:
        multimetro.write("TRIG:COUN 1")
        multimetro.write("INIT:CONT OFF")
    leituras = []
    try:
            time.sleep(0.5)
    return leituras
Offline Rubo
« Reply #6 on: July 25, 2025, 03:38:38 pm »
Agreed. You should use "FETCh?" instead if you want continuous readings.
« Last Edit: July 25, 2025, 03:41:21 pm by Rubo »
Offline KungFuJosh
  • TEAS is real.
« Reply #7 on: July 25, 2025, 04:16:34 pm »
I've not tried with the Keithley, but using the same SCPI logic on the CNT-90 and 53132A, READ? was always better, and to use FETC? required initiating the measurement with 2 or 3 commands. FETC? only gave the last measurement and didn't always give a new reading even in continuous mode. 🤷
Offline Rubo
« Reply #8 on: July 25, 2025, 04:57:20 pm »
Now it's really confusing.
In HOIKI programming manual it says that READ initiates a single trigger event and returns the reult of said event, so you have use it with INIT:CONT OFF
FETCh does not initiate a trigger event and simply return whatever the last triggered reading was, so you have use it to display results with INIT:CONT ON
In Keysight's programming guide it gives similar explanation, but says you can use READ with INIT:CONT ON
 |O  |O  |O
Offline KungFuJosh
  • TEAS is real.
« Reply #9 on: July 25, 2025, 05:24:14 pm »
I think maybe Hioki has it backwards there since I'm guessing Keithley and Keysight may agree?
Offline TizianoHV
    • My Website
« Reply #10 on: July 25, 2025, 10:31:33 pm »
I have a Keithley 2790 (2000 family) and an Agilent 34970A (similar to 34401A).
Keithleys are a bit pickier with their trigger model (you need to "manually" disable INIT cont).
The Agilent will enter automatically "remote" mode (so you don't need to send INIT:CONT OFF).
  • Country: pt
« Reply #11 on: July 30, 2025, 01:57:10 pm »
First of all, I thank you for all the help you gave me. I see that there are some differences between equipment in terms of SCPI commands. I managed to get it working without error, here is the working code. Does it make sense to stay like this?First of all, I thank you for all the help you gave me. I see that there are some differences between equipment in terms of SCPI commands. I managed to get it working without error, here is the working code. Does it make sense?
    def enviar(cmd):
        try:
            multimetro.write(cmd)
            erro = multimetro.query("SYST:ERR?")
            if not erro.startswith("0"):
                print(f"[ERRO SCPI] Comando: {cmd} → {erro.strip()}")
            print(f"[EXCEÇÃO PYTHON] ao enviar '{cmd}': {e}")
    try:
        enviar('FUNCtion "RESistance"')
        enviar("SENS:RES:RANG:AUTO ON")
        enviar("SENS:RES:NPLC 1")
        print("[Keithley DMM6500] Modo resistência configurado.")
        print(f"[ERRO] na configuração: {e}")
Now I have another challenge which is to read in diode mode |O |O
« Last Edit: July 30, 2025, 02:00:38 pm by magama »
Offline Rubo
« Reply #12 on: July 30, 2025, 02:51:07 pm »
If it's working as expected, sure. I would still issue *RST & *CLS in the beginning of the code, before -> enviar("SENS:RES:RANG:AUTO ON")
Have fun  :popcorn:
Offline TizianoHV
    • My Website
« Reply #13 on: August 18, 2025, 10:07:07 am »
When you send ""SYST:ERR?" you'll get only the first error (FIFO). If you want to read all errors you should use a while loop (easy fix).