Author Topic: Fluke 8840a fault?  (Read 691 times)
Offline reddish75Topic starter
Fluke 8840a fault?
« on: August 30, 2026, 08:54:49 pm »
I've recently bought a used possibly faulty? Fluke 8840a bench meter which has the Ac measurement and IEEE interface as well, so again decent enough spec'd meter I feel.
Anyway the fault I think is that the meter dosent zero, with or without probes attached and even when the probes are shorted together,  offset dosent seem to do much.
Checking main psu voltages they are in spec.
Any help where to look Inwould be grateful
Re: Fluke 8840a fault?
« Reply #1 on: August 30, 2026, 09:16:10 pm »
With probes shortest together
Online zike
« Reply #2 on: August 30, 2026, 09:47:26 pm »
Check the front/rear terminal switch.
Offline bdunham7
« Reply #3 on: August 30, 2026, 10:04:13 pm »
Get the manual, do the self-test routines.  Gotta wonder why V and ma are both showing at the same time.
« Reply #4 on: August 31, 2026, 08:19:51 am »
Yeah, that's very strange. Best hope it's not one of the input MUX ICs, those are pretty much unobtainium.
Online Sorama
« Reply #5 on: August 31, 2026, 08:47:27 am »
segmens are sometimes missing too.
Might be 'just' a problem with the display itself?
Offline bdunham7
« Reply #6 on: August 31, 2026, 02:01:59 pm »
I doubt the problem is with the display itself, but I suppose anything is possible.  But running the self-tests would help a lot.  This looks like an "interesting" problem.  :)
« Reply #7 on: August 31, 2026, 07:17:31 pm »
I've ran the self tests and it passes them all, I'm away for a few days now but shall run them again and report back
Offline bdunham7
« Reply #8 on: August 31, 2026, 10:30:43 pm »
If it passes the self-tests, then you likely have an issue with U202 or U212 (the Z8 microprocessor) or its ROM/EPROM or some closely related circuit in that section.  The only thing that doesn't line up with that is that max negative reading with the inputs shorted.  Try each range shorted and open to see what the display does, paying attention to whether the faults are completely consistent or sort of random.
Offline BillyO
« Reply #9 on: August 31, 2026, 11:14:18 pm »
If you have a oscilloscope, check for ripple on the DC supplies.
Bill