Hi all,
I own some Tektronix active probes (ADA400A and P6247) that still use the TekProbe interface [1]. I have quite a hard time to get them running with my Siglent SDS5000X via the official Siglent TPA10 TekProbe adapter [3]. In particular, I am unable to do any offset adjustment, and I have a suspicion that my adapter is defective.
I realized I went into way too much detail, so here is the short version:
Could someone who owns EITHER a Tektronix Oscilloscope with a TekProbe interface OR a Siglent Oscilloscope with a TPA10 adapter please measure the voltage present at the offset pad of the TekProbe interface?
- Does the voltage change at all when you rotate the offset button?
- What is the voltage range when adjusting the channel position to min/null/max?
-------------Rambling and lengthy debugging below----------------
Info: Both probes support manipulating the DC offset. The ADA400A supports manual offset adjustment, while the P6247 offers DC Reject.
Info2: The mechanism by which the TPA 10 generates its offset is the following (provided by Siglent):
The mechanism of the scope controlling the offset over TPA10 is that the scope programs an offset DAC in the TPA10, and the TPA10 outputs the offset to the Tek's probe thru the TekProbe interface level II. If the offset is correctly added to the front end of the probe, then it can be captured and displayed on the scope.
What I tried:
ADA400A:1) When DC offset is enabled on the probe:
- The ADA400A allows manual adjustment on the probe itself.
- The offset control of the scope does not work.
2) When DC offset is disabled on the probe:
- The ADA400A no longer controls the offset, but neither does the scope.
- The offset control of the scope does not work.
I think this expected due to the limited interface on this probe; there are only 4 wires going from the probe to the TekProbe connector (+/-15V, +5V, attenuation ID), so the offset voltage is ignored.
P6247:1) When DC reject is enabled on the probe:
- The P6247 rejects DC and shows correct signals.
- The offset control of the scope does not work.
2) When DC reject is disabled on the probe:
- The P6247 shows a constant DC offset of about 1.2V, which I assume means it is crashing into the V+ rail of the input transistor.
- When applying the input calibrator of my scope to the probe input, i can see a faint square wave riding on top of the 1,2V DC offset.
- The offset control of the scope does not work.
To find out what is going wrong, I measured the offset pad in the tekprobe adapter. It is always sitting at 2.5V, no matter how I adjust the offset. That seems strange to me, since for the 1103 Tekprobe power supply, Tek adjvertises a +-1V offset range [2]:
Outputs
Pin 1: n.c.
Pin 2: n.c.
Pin 3: +5 V.
Pin 4: +15 V.
Pin 5: Offset ±1 V.
Pin 6: –5 V.
Pin 7: –15 V.
The adapter works the same way (taken from [3]):
Specifications
Bandwidth 4 GHz
Power Supplies +15 V (±2%, 100 mA),
-15 V (±2%, 100 mA),
+5 V (±2%, 200 mA),
-5 V (±2%, 200 mA)
Offset Voltage -1.2 V ~ +1.2 V to the probe. Multiply it by the attenuation factor of the probe to calculate the system offset
range.
Now if the offset works the way I think it works, a fault where it is 2.5V all the time would surely cause the probe to crash into Vmax.
Could someone who owns a Tek Scope with a TekProbe interface please measure the offset pin? What is the voltage range? Does it change when you rotate the offset button?
If someone has a Siglent TPA10: Could you verify if your adapter behaves the same way? What is the voltage range? Does it change when you rotate the offset button?
Thanks a lot,
BR Andreas
[1]
https://www.tek.com/en/support/faqs/what-tekprobe-interface[2]
https://www.tek.com/en/datasheet/probe-power-supply-1103[3]
https://siglentna.com/wp-content/uploads/2021/10/LPA10_TPA10_Datasheet_DS060APA_E01A.pdf