My unit behaves much the same way in regards to sensitivity, except that it reads significantly high in the low range with an input. If I ground the input on the .0003 range the meter reads about 2 minor ticks above zero...I believe my problem is one of scale and not offset.
Could my setup be part of the problem? I'm using a restored 200CD with wire leads and to make it usable in the microvolts I had to place a voltage divider at the input of the 334 consisting of a 10KR in series and a 1R across the input and ground. To read the input voltage I have a 34401A clipped across the 1R.
To clarify my previous post I was going through the troubleshooting guide figure 7-9b in the manual. I think the positions of the negative voltages are flipped top-bottom.
(1) I verified the front input and the meter circuit input agree. I can adjust amplitude and the readings stay within several uV of each other. For the following checks and to read the meter the multimeter leads are removed from pin 6 and 15 (ground). On .0003 range meter is off scale above a front input of 150uV.
(2) Input signal returned to 300uV. Q7 output looks about right...
(3) Amplitude is too high.
(4) Also too high...
The bizarre thing is that on other ranges the waveform voltages are correct at full deflection and it reads accurately, and the error affects both .0003 and .001 ranges. The periodic noise can't be normal. It's worth noting that it looks like A2Q1 was desoldered for either testing or replacement.
It seems as though I cannot post images in-line...?