I just bought an HP 3456A multimeter that was classified as non-operational. The symptom of its non-operation is that the unit itself turns on, begins to sample as indicated by the sample rate annunciator LED flashing, but the rest of the display stays off.
So, first thing first, I checked all of the voltages on the power boards and they all seem to check out. There don't appear to be any bad electrolytics either. Then, I went for the A4 board (Keyboard scan logic, display and annunciator). Voltages on that board seem fine. I see level changes on U29 and U30. I see a clock feeding U28 (an octal latch). The 5V/Gnd test points on the board measure 5 V. Data seems to be arriving at the D0-D7 pins. Nothing looks wrong or suspicious. I decided to disconnect the ribbon connecting from J4 to the A2 display/button board. I examined the ribbon and it is in excellent condition by the way with no damage. I pulled out the A4 board itself, cleaned that J4 connector on the board to which the display/keyboard ribbon connects. Made sure when I plugged the board in that everything is snug - both the ribbon that connects to J4 as well as the two rows of pins that hold the board.
However, the problem persists - nothing on the display lights up, not the 7-segment LEDs and not the LEDs next to buttons. The funny thing is that the keys seem to work and perform their proper functions, although I only have the sample rate annunciator LED to tell me that since it's the only darn thing that lights up. None of the 7-segment LEDs nor the LEDs near the buttons light when I press the buttons near them. The buttons themselves seem to work, because they have an effect on the flashing of the sample rate annunciator LED - an effect that seems reasonable for their intended purpose.
The funny part about all of this is that the button LEDs, the annunciator 7-segment display and the other 7-segment displays are all being driven by different seemingly functioning driver chips (I checked input voltages on them for +5/Gnd, by the way) on the A4 board. Yet, none of the LEDs on A2 (Keyboard, Displays and Annunciators) light up other than that one sample rate annunciator LED which flashes in utter defiance of the rest of its neighbors!
I also tried pressing the TEST button which is supposed to light up all of the LEDs, at least on the 7-segment panel, but nothing lights up even though the test seems to function and cycles like it should, which I hear from relays closing/opening, indicating (probably) successful tests cycling.
I am at wit's end. The A2 board with the actual LEDs and seven segments is very hard to get at without a lot of diassembly so its difficult for me to get probes in there, but I still managed to test some of the 7-segment pins and see voltages reaching them (both, several volts and a few hundred millivorts north of ground).
So, what could be causing all of the LEDs and LED segments but one which are driven by multiple driver chips not to work? It seems statistically impossible that all but one of the LEDs just decided to die.
I know many of you guys have HP 3456A troubleshooting experience, so thanks for any and all advice, I really appreciate it!