One thing that makes it feel smaller too is that the trigger doesn't seem to fire when the waveform is zoomed out too much. With a 1.4Vpp sine wave at 10V/div with a DC rising edge trigger at 0V it doesn't fire. The same settings to fire the trigger on the Tek scope we're evaluating. Any ideas what's going on here?
First let us get the facts right: Trigger sensitivity is universally specified as divisions. A 1.4 Vpp signal at 10 V/div fills just 0.14 div.
Now let's see,what Tektronix has to say on that matter:
https://www.tek.com/en/support/faqs/what-are-reasonable-trigger-thresholds-how-small-signal-will-trigger-oscilloscopeHere we learn, that "Most Tektronix oscilloscopes will trigger on a signal of 0.35 divisions at up to 50 or 100 MHz". This is a pretty high sensitivity indeed, yet it's only guaranteed at lower frequencies, up to 100 MHz. Most importantly, that minimum level is still 3.5 Vpp, hence 2.5 times the 1.4 Vpp signal.
By contrast, the Siglent SDS3000X HD has a trigger sensitivity of 0.52 div over the full bandwidth. So yes, it is a bit less sensitive, but at the same time also more noise tolerant and not bandwidth limited. The special "Noise Reject" setting increases the trigger threshold to even higher 0.66 div.
The question remains, why someone would want to dsiplay a signal trace with such a tiny amplitude - and then trigger on it, on top of that?
In general, vertical sensitivity should be chosen to maximize the signal without clipping. Just like an analog meter, where you try to get the measurement in the upper third of the scale.
With 0.14 div, only ~1.7 % of the ADC range is used. In an 8-bit system, this reduces the acquisition to 2 bit resolution - it's still 6 bits in a 12-bit system. Either way, such a tiny signal is not a good trigger source, especially nowadays, where fully digital trigger systems are almost universally used.
A trigger signal should always be as much full scale as possible. If the signal is of no interest otherwise, then making it tiny by reducing the vertical gain is the wrong approach. One can just hide the trace or use the external trigger input instead.