Ah, only got it now. You talk about the resistor on instrument side.
HP3547a has R548 on the Main PCB supplying 5V to power the 5V side of the isolator. I had put a note for it on the PCBs back side.
Now for your instrument it's a different situation with 2 Rs it seems
Edit: as Ko4bb.com has an issue and I don't have access to the service manual, do you remember the reference designstoes for those 2 resistors? Would like to put those too on the PCB silkscreen.
Yes there are 2 Rs - one is on the +5V rail and one is on the Ground rail. These are 270 ohm each and the are designated R205 and R206 . I replaced one of them with 100 ohm from the ground side. The isolator chip current is definitely the issue as it is specified between 1mA and 3 mA quiescent current. Looking at the operating frequency of these displays below 100khz we can assume idle current and working current to be the same. That would create a voltage drop on the power rail for the ISO chip.. I guess I got lucky that it was not a problem for the other Meters or Power supplies that I installed the OLED. I think these instruments have 1k resistors.
I would say that because of the operating spec we should keep the resistors low value - 220 ohms maybe max per display. That will limit any drop to 0.66V . I am guessing these offer some protection from plugging in the display backwards in the original design so we should probably respect that. But after deeper dive - if you connect the display cable backwards I see the +5V will go to a No connect and Ground will go to INA input. So in theory no damage would be done. My instrument also as a keyboard matrix with the same connector type - more likely the resistor is to protect from plugging in the keyboard into the display port . That could destroy the carbon pads on the keyboard - maybe something HP was trying to protect against? A 220 ohms resistor would allow a short circuit current of 23mA. A little high for carbon switch but will have to be OK.
These might have had dual purpose - one a protection purpose - 2nd function RC low pass to prevent back feeding noise.
Either way 100 ohms each in HP3852A will drop max 1.2V (0.4,,,1.2V) and seems to work just fine. I will measure tonight the actual voltage drop but chip to chip it a vary. So if you have issue with your display this might be the problem - too high dropper resistors on the power rails.
Kawal